Re: Oil objective maximum area/distance for long-term timelapse imaging

Posted by Jason Miller on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Oil-objective-maximum-area-distance-for-long-term-timelapse-imaging-tp7584835p7584843.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We tried to solve this problem by using water instead of oil, acknowledging
that we lost the increased NA. We were using an inverted setup, and similar
to Sylvie, we machined a cap that goes around the objective with a drip
setup - so water drips into this cap, which encases the objective, and
keeps water in the cap at all times. You have to empirically test what drip
rate you need to counter evaporation. We found that the viscous water
substitutes (e.g. ultrasound gel, gonisol, etc.) did not have enough
capillary action to stay sealed between the objective and the bottom of the
plate. But water did. This also allows for very fast imaging - you don't
have to slow down the speed with which the stage travels between wells or
within wells at all.

Regards-
Jason

-------

Jason Miller, MD, PhD

University of Michigan Kellogg Eye Center


Home address:

117 Worden Ave

Ann Arbor, MI 48103

Cell: (415) 225-2134<tel:%28415%29%20225-2134>

E-mail: *[hidden email]<mailto:[hidden email]> <[hidden email]<mailto:[hidden email]>>*


"Forgiveness does not change the past, but it does enlarge the future." -
Paul Boese

On Wed, Mar 2, 2016 at 12:05 PM, Sylvie Le Guyader <[hidden email]<mailto:[hidden email]>>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your posting.
> *****
>
> I forgot to write that what the magic oil device does is that it refills
> the immersion oil on the objectuf lense. :-)
>
> I agree that reducing the stage speed helps but we never got to a fully
> reliable system. Might take magic hands... Sigh!
>
> Ann. What do you mean by 'it did not keep'? Did it bleach? The GattaQuant
> website specifically says that it virtually doesn't bleach! :-(
>
> Med vänlig hälsning / Best regards
>
> Sylvie
>
> @@@@@@@@@@@@@@@@@@@@@@@@
>
> Sylvie Le Guyader, PhD
> Live Cell Imaging Unit Manager
> Karolinska Institutet- Bionut Dpt
> Hälsovägen 7,
> Novum, G lift, floor 6
> 14157 Huddinge
> Sweden
> mobile: +46 (0) 73 733 5008<tel:%2B46%20%280%29%2073%20733%205008>
> office: +46 (0) 8 5248 1107<tel:%2B46%20%280%29%208%205248%201107>
> LCI website
>
>
> ---- Smith, Benjamin E. wrote ----
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your posting.
> *****
>
> I can also testify to the benefit of greatly reducing the stage velocity
> during long-term live imaging.  We have done imaging of up to 48-hours,
> and having the stage crawl along allows the immersion fluid to stay wetted
> to the objective, and also eliminates any sample motion.
>
> One other consideration we've done is to setup the run early in the
> morning so you can check up on it throughout the day.  That way, you can
> also add immersion fluid as needed.  Normally, you will find that towards
> the beginning of the run you may need to add a little more to the
> objective, but after about 3-4 hours everything stabilizes, and the system
> is good to go.
>
> I also make sure that there is a sufficient pause between scans so that I
> have enough time to adjust things and still start the next scan at exactly
> the next time interval.
>
> Good luck,
>    Ben Smith
>
> ________________________________________
> From: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> on
> behalf of Emmanuel Levy <[hidden email]<mailto:[hidden email]>>
> Sent: Wednesday, March 2, 2016 8:46 AM
> To: [hidden email]<mailto:[hidden email]>
> Subject: Re: Oil objective maximum area/distance for long-term timelapse
> imaging
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your posting.
> *****
>
> It may also help to use oil with high viscosity index so it remains spread
> on the glass.
> All the best,
> Emmanuel
>
>
> On 2 March 2016 at 16:18, Gary Laevsky <[hidden email]<mailto:[hidden email]>> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your
> posting.
> > *****
> >
> > We regularly image the 4 center wells (about 50 total points) of a
> Labtek 8
> > well chambered coverslip for 8-16 hours at 37c using PFS on a TiE.
> >
> > First we put the plate on, one drop of oil on the 60X or 100X objective,
> > and drive to all four wells.
> >
> > Then we take the plate off, and add one more drop of oil, and we're good
> to
> > go.
> >
> > We also slow the stage speed way down so PFS can follow the glass.
> >
> > Hope this helps.
> >
> > Gary
> >
> > On Wed, Mar 2, 2016 at 7:49 AM, Sylvie Le Guyader
> <[hidden email]<mailto:[hidden email]>
> > >
> > wrote:
> >
> > > Hi Ken
> > >
> > >
> > >
> > > My experience: imaging 2 cells 1.7mm apart (thickness of the well wall
> in
> > > the MW plates we were using then) every 10 min for 12h with a 60x oil
> > > objective and a hardware autofocus at 37 deg resulted in a high risk
> > (about
> > > 25% of the cases?) of losing focus. There go years of sweat and tears...
> > >
> > >
> > >
> > > We were lucky to get to test a super device manufactured by EMBLEM,
> the
> > > commercial side of EMBL. You send your objective. They custom make a
> cap
> > > and send you a kit containing cap, tubes, pump and instructions. It is
> a
> > > bit clumsy because you need to remove all other objectives and you
> need
> > to
> > > be careful not to put oil all over the place but it works a charm. I
> have
> > > imaged 10 positions in each well over 15 wells every 10 min for 15 h
> with
> > > this device without losing focus. :) You could contact them<
> > >
> http://www.embl-em.de/articles.php?lang=de&hid=4&Cat=Technology_Offers>
> > > to ask if they commercialize it.
> > >
> > >
> > >
> > > Med vänlig hälsning / Best regards
> > >
> > >
> > >
> > > Sylvie
> > >
> > >
> > >
> > > @@@@@@@@@@@@@@@@@@@@@@@@
> > >
> > > Sylvie Le Guyader, PhD
> > >
> > > Live Cell Imaging Facility Manager
> > >
> > > Karolinska Institutet- Bionut Dpt
> > >
> > > Hälsovägen 7,
> > >
> > > Novum, G lift, floor 6
> > >
> > > 14157 Huddinge
> > >
> > > Sweden
> > >
> > > mobile: +46 (0) 73 733 5008<tel:%2B46%20%280%29%2073%20733%205008>
> > >
> > > office: +46 (0) 08-524 811 72 New number!
> > >
> > > LCI website
> > >
> > >
> > >
> > >
> > >
> > > -----Original Message-----
> > > From: Confocal Microscopy List
> [mailto:[hidden email]<mailto:[hidden email]>]
> > > On Behalf Of Kenneth Chen
> > > Sent: den 2 mars 2016 13:22
> > > To: [hidden email]<mailto:[hidden email]>
> > > Subject: Oil objective maximum area/distance for long-term timelapse
> > > imaging
> > >
> > >
> > >
> > > *****
> > >
> > > To join, leave or search the confocal microscopy listserv, go to:
> > >
> > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > >
> > > Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your
> > posting.
> > >
> > > *****
> > >
> > >
> > >
> > > Hi all,
> > >
> > >
> > >
> > > We're in the process of designing microfluidic devices for multi-day
> > > time-lapse imaging. In the past we've had some trouble with image
> quality
> > > for very large devices deteriorating due to the loss/thinness of the
> > > immersion oil layer between sample and objective when very large areas
> > are
> > > scanned. We have only used traditional oil objectives (no water or
> > >
> > > glycerol) Does anyone have any strategies to mitigate this or rules of
> > > thumb about "safe" maximum sampling areas or distances to travel?
> > >
> > >
> > >
> > > Thanks,
> > >
> > > Ken Chen
> > >
> >
> >
> >
> > --
> > Best,
> >
> > Gary Laevsky, Ph.D.
> > Confocal Imaging Facility Manager
> > Dept. of Molecular Biology
> > Washington Rd.
> > Princeton University
> > Princeton, New Jersey, 08544-1014
> > (O) 609 258 5432<tel:609%20258%205432>
> > (C) 508 507 1310<tel:508%20507%201310>
> >
>
**********************************************************
Electronic Mail is not secure, may not be read every day, and should not be used for urgent or sensitive issues