Hi, Kyle
The general rule of thumb is that the NA on the condenser should meet
or exceed that of the objective.
If you are using oil immersion objectives, ideally, to achieve that
goal, you should use an oil immersion condenser, otherwise you are
limited to an NA of 0.9.
Also, remember that the aperture iris in the condenser adjusts the
condenser's WORKING numerical aperture. Just because the condenser
is marked 1.4 NA doesn't mean that, in a practical experiment, it
will be operating at 1.4. I follow the guidelines set down by Frits
Zernicke (inventor of Phase contrast): gently close the aperture
iris to the "Oomph" position: that delicate balance between
sufficient edge definition and optimum resolution (Yes, the condenser
does contribute to resolution).
As for planning for growth:
You might want to invest in a turret condenser early on. That will
give you the option to add those other contrast techniques as you
grow into them.
And just one more reminder, specifically regarding DIC:
If you are going to use plastic vessels (petri dishes, multi-well
plates, growth flasks), use Hoffman Modulation Contrast instead of
DIC. DIC uses polarized light. The plastic will affectt the shear
and cause effects that will be hard to interpret. Some HMC set-ups
do use pol to control the width of the slit in the condenser, but all
of that is on the incoming side of the sample and will not be
affected by plastic containers.
Good hunting!
Barbara Foster, President & Chief Consultant
Microscopy/Microscopy Education ... "Education, not Training"
7101 Royal Glen Trail, Suite A - McKinney, TX 75070 - P: 972-924-5310
www.MicroscopyEducation.com
Microscopy/Microscopy Education is a division of The Microscopy &
Imaging Place, Inc.
NEW! Getting involved in Raman or FTIR?
MME is now offering courses in these areas specifically for microscopists!
Now scheduling courses through the mid 2016. We can customize a
course on nearly any topic, from fluorescence to confocal to image
analysis to SEM/TEM.
Call today for a free training evaluation.
At 08:36 AM 3/17/2016, Kyle Michael Douglass wrote:
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>Post images on
http://www.imgur.com and include the link in your posting.
>*****
>
>Hello listers,
>
>
>I have a couple of questions about condensers for you. I'd like to
>do some transmitted light imaging in an inverted microscope using
>high magnification, oil-immersion objectives. For the moment, I
>don't need to do anything other than brightfield with a high power
>LED light source. It might be nice to do phase contrast or DIC in
>the future, but I don't need it now.
>
>
>My questions are:
>
>
>1) What are the rules of thumb for matching a brightfield condenser
>to an objective? I won't be using anything but oil immersion
>objectives with NA's greater than 1.4.
>
>
>2) If I do want to do phase contrast or DIC in the future, should I
>put special consideration into the condenser lens selection now? I
>imagine the condenser NA will determine what phase contrast rings I
>can use, but does it impact DIC?
>
>
>Thanks!
>
>Kyle
>
>
>Dr. Kyle M. Douglass
>Post-doctoral Researcher
>EPFL - The Laboratory of Experimental Biophysics
>
http://leb.epfl.ch/>
http://kmdouglass.github.io