Thanks for the feedback, Barbara. It is very helpful.
greater than or equal to the objective NA. Can you offer some physical
consists of two parts. One part is the transmitted light that is not
scattered by the sample. The other part is the light scattered by the
sample. If the condenser's working NA is smaller than the objective's,
objective's back focal plane. However, the light scattered by the sample
NA requirements of the condenser and objective. Does this make sense?
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi, Kyle
>
> The general rule of thumb is that the NA on the condenser should meet
> or exceed that of the objective.
>
> If you are using oil immersion objectives, ideally, to achieve that
> goal, you should use an oil immersion condenser, otherwise you are
> limited to an NA of 0.9.
>
> Also, remember that the aperture iris in the condenser adjusts the
> condenser's WORKING numerical aperture. Just because the condenser
> is marked 1.4 NA doesn't mean that, in a practical experiment, it will
> be operating at 1.4. I follow the guidelines set down by Frits
> Zernicke (inventor of Phase contrast): gently close the aperture iris
> to the "Oomph" position: that delicate balance between sufficient edge
> definition and optimum resolution (Yes, the condenser does contribute
> to resolution).
>
> As for planning for growth:
> You might want to invest in a turret condenser early on. That will
> give you the option to add those other contrast techniques as you grow
> into them.
>
> And just one more reminder, specifically regarding DIC:
> If you are going to use plastic vessels (petri dishes, multi-well
> plates, growth flasks), use Hoffman Modulation Contrast instead of
> DIC. DIC uses polarized light. The plastic will affectt the shear
> and cause effects that will be hard to interpret. Some HMC set-ups do
> use pol to control the width of the slit in the condenser, but all of
> that is on the incoming side of the sample and will not be affected by
> plastic containers.
>
> Good hunting!
> Barbara Foster, President & Chief Consultant
> Microscopy/Microscopy Education ... "Education, not Training"
> 7101 Royal Glen Trail, Suite A - McKinney, TX 75070 - P: 972-924-5310
> www.MicroscopyEducation.com
>
> Microscopy/Microscopy Education is a division of The Microscopy &
> Imaging Place, Inc.
>
>
> NEW! Getting involved in Raman or FTIR?
> MME is now offering courses in these areas specifically for
> microscopists!
> Now scheduling courses through the mid 2016. We can customize a
> course on nearly any topic, from fluorescence to confocal to image
> analysis to SEM/TEM.
> Call today for a free training evaluation.
>
>
>
>
> At 08:36 AM 3/17/2016, Kyle Michael Douglass wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> Post images on
http://www.imgur.com and include the link in your
>> posting.
>> *****
>>
>> Hello listers,
>>
>>
>> I have a couple of questions about condensers for you. I'd like to do
>> some transmitted light imaging in an inverted microscope using high
>> magnification, oil-immersion objectives. For the moment, I don't need
>> to do anything other than brightfield with a high power LED light
>> source. It might be nice to do phase contrast or DIC in the future,
>> but I don't need it now.
>>
>>
>> My questions are:
>>
>>
>> 1) What are the rules of thumb for matching a brightfield condenser
>> to an objective? I won't be using anything but oil immersion
>> objectives with NA's greater than 1.4.
>>
>>
>> 2) If I do want to do phase contrast or DIC in the future, should I
>> put special consideration into the condenser lens selection now? I
>> imagine the condenser NA will determine what phase contrast rings I
>> can use, but does it impact DIC?
>>
>>
>> Thanks!
>>
>> Kyle
>>
>>
>> Dr. Kyle M. Douglass
>> Post-doctoral Researcher
>> EPFL - The Laboratory of Experimental Biophysics
>>
http://leb.epfl.ch/>>
http://kmdouglass.github.ioKyle M. Douglass, PhD