Hi Lili,

With the scope fully off try to clean round metal leads on the bottom surface of sample
holder and corresponding connections on the microscope frame with some alcohol. We had a
similar problem due to oil spills and simple cleaning helped. Also check if you have the
same issue with other sample holder(s) if you have them. Hope you will bring it back to life!