> Hi Rafael,
>
> How thick do you think the sample is?
> The pinhole is ~3 Airy Units, which means the confocal section thickness is ~20 microns.
> So if you had great SNR in the confocal image, you may not see a difference in the two images. By eye they look very similar, except for the reduced # of photons in the confocal image.
>
> The formula can be found here (Eq. 5):
>
http://zeiss-campus.magnet.fsu.edu/articles/spinningdisk/introduction.html>
> I have it in an Excel spreadsheet if anyone is interested (but can't vouch for its complete accuracy).
>
> Cheers,
> Jeff
>
>
> -----Original Message-----
> From: Rafael Jaimes III [mailto:
[hidden email]]
> Sent: Friday, May 20, 2016 4:54 PM
> To:
[hidden email]
> Subject: Re: widefield getting better images than spinning disk
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Thanks Sam,
>
>
http://imgur.com/a/KQysd>
> Let me know if the above album works for you. I uploaded two images: one in widefield and one in spinning disk confocal. All things else are equal. For this example I used a fixed sample with 80 ms exposure. The image is at 10x, 0.45 NA with confocal pinhole of 40 um. Is this amount of image degradation to be expected? I also tried turning up the light source intensity significantly, with not much difference.
>
> Rafael
>
>> On Fri, May 20, 2016 at 12:57 PM, Sam Lord <
[hidden email]> wrote:
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> Post images on
http://www.imgur.com and include the link in your posting.
>> *****
>>
>> On Fri, 20 May 2016 12:46:37 -0400, Rafael Jaimes III
>> <
[hidden email]
>> wrote:
>>> 1. Live neonatal rat cardiomyocyte-fibroblast layers stained with
>>> Fluo-4 for calcium. They are field stimulated and imaged >100 fps.
>>
>> At 10 ms exposure time, I doubt you're getting enough fluorescence
>> hitting the camera to get a sufficient signal to noise. Wide field is
>> always much brighter than confocal because both the excitation light
>> and the emission intensity is reduced going through the pinholes.
>>
>> Why don't you send images of wide field vs confocal so we can see what
>> you mean by "better."
>>