dear Jeff,
collection efficiency CF/WF is 5% to 6% in our systems with 70um pinholes
(fill factor 8.2%). 4.6-4.8% with 60um (fill factor 5.8%). about 3-3.5% with
40um pinholes (fill factor 4.2%). values measured at 500nm.
hence, my proposal to double check for proper system alignment (pinholes to
detector focusing and excitation).
regards
Andrea
Andrea Latini
President
CrestOptics Srl
On Sun, 22 May 2016 18:03:48 +0000, Reece, Jeff (NIH/NIDDK) [C]
<
[hidden email]> wrote:
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>Post images on
http://www.imgur.com and include the link in your posting.
>*****
>
>I was wondering how long it would take before Guy responded. There have
been discussions on the listserv in years past about signal as a function of
confocal pinhole, and Guy has always stressed this basic concept. I can't
find those previous discussions, but there is a nice graph here that
illustrates the concept (scroll down to Figure 6, the graph on the right):
>
http://www.leica-microsystems.com/science-lab/super-resolution-on-a-heuristic-point-of-view-about-the-resolution-of-a-light-microscope/(disclaimer: this is the only place I could find it on the web; thanks
Leica) which makes sense if you look at Figure 3 on the same page, just
visually estimating how much of the total light from the psf must be inside
the Airy Disk.
>So, another way to state the same concept: only 16% of in-focus light is
thrown away when the pinhole is ~1 AU (i.e. "confocal" in the general sense).
>If anyone has the original references that show the graph of integrated
intensity vs AU, or another place it might be on the web for free, I would
be interested. Perhaps those links are with the previous discussion on the
listserv that I can't find.
>
>With the 10x/0.45 lens, and the pinhole at ~3 AU, you are collecting more
like 94% of the in-focus light.
>And since the FWHM z-resolution is ~20 microns for that pinhole and lens
(assuming 500nm as the emission wavelength), then you are collecting ~47% of
the out-of-focus light that originates from 10 microns away from the focal
plane.
>
>My understanding of the Crest Disk, from the info I see on their web site,
is that throughput on the excitation side is only 1.6%, compared to
widefield illumination (disk pulled out). So if you increase the camera
exposure time by a factor of ~70 on the confocal image, the signal levels of
the in-focus portion should be similar to the widefield image, if everything
is aligned properly. Theoretically.
>
>Hope that helps.
>Cheers,
>Jeff
>
>
>-----Original Message-----
>From: Guy Cox [mailto:
[hidden email]]
>Sent: Sunday, May 22, 2016 1:55 AM
>To:
[hidden email]
>Subject: Re: widefield getting better images than spinning disk
>
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>Post images on
http://www.imgur.com and include the link in your posting.
>*****
>
>Repeat 100 times:
>
>CONFOCAL ONLY THROWS OUT
>OUT OF FOCUS LIGHT!
>
>The reason spinning disk systems lose light is to attain speed, and has
nothing to do with the confocal principle.
>
> Guy
>
>Guy Cox, Honorary Associate Professor
>School of Medical Sciences
>
>Australian Centre for Microscopy and Microanalysis, Madsen, F09, University
of Sydney, NSW 2006
>
>-----Original Message-----
>From: Confocal Microscopy List [mailto:
[hidden email]] On
Behalf Of Sam Lord
exposure it takes to get a good image. If the image gets bright but the
contrast still looks worse than wide field, then maybe there's an alignment
issue. But I suspect your sample is just too dim to image with confocal at
10 ms per frame. Not many samples are bright enough for that.
>Remember that confocal is designed to throw out light in order to improve
optical slicing.