Hi Steve:
We do it for labs when we visit, or you can try it yourself.

It can be done in about 15 minutes. The thing to be careful with is not to remove the screws completely and not to scratch the light absorbing surface of the retaining spring and light shield.
Having the right screw driver helps;  PH000 x 40 (Wiha) works. The shields are available but expensive and delicate. The newer model CSUX and W1 systems are easier.

Be careful to insert the dichroic all the way up to the seat limit and have the dichroic oriented correctly with the coated side facing the microscope. Also avoid doing this in dusty rooms and do not use canned air to blow off dust,... better done with squeeze bulb with filtered air intake.

Not rocket science, but does take some care.

Best

George


George Peeters, M.D., M.S.
Pres. Solamere Technology Group inc.
1515 Military Way
Salt Lake City, UT.  84103

Cell    801 232-6911





On May 24, 2016, at 11:10 AM, Bunnell, Stephen C. <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> Has anyone had any experience with installing a dichroic into the static mount of a Yokogawa CSU-10 spinning disc unit? I have three dichroic mounts and need to swap out one dichroic for an updated version. Yokogawa no longer supports the CSU-10, so any advice would be appreciated.
>