Have you tried redirecting the blue and green channels to the detector that is usually “red”?
after ~28,000 hrs on our FV-1000, the “green” channel started dropping out, sometimes entirely, other times with dark bands in the scanned frame. The problem persisted when routed to other detectors. The culprit was the AOTF.
> On Jun 1, 2016, at 10:35 AM, jens rietdorf <
[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear colleagues,
>
> thank you all for your help on the issue, it turned out a shutter inside
> the scanhead was blocked. This problem was relatively rapidly addressed by
> the service of Olympus Brasil and Olympus Americas. Very helpful input for
> the differential diagnosis came from Farid Jalali of Olympus Canada.
>
> A new (maybe related) issue occurred now, we experience very low
> sensitivity on detector channels 1&2,
> channel 3 operates normally. the effect appears to be stronger in the
> blue/green part of the spectrum than for the red.
> The FV diagnostic tool does not indicate any abnormalities.
>
> Again, all ideas and suggestions are very welcome.
>
> Kind regards, Jens
>
> Dr. Jens Rietdorf, visiting scientist @ center for technological
> development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de Janeiro
> Brasil.
>
> On Mon, Apr 4, 2016 at 6:39 PM, jens rietdorf <
[hidden email]> wrote:
>
>> Listers,
>>
>> following the installation of a new air condition, we are not succeeding
>> to get any light after the scanhead of our Olympus FluoView1000 CLSM. End
>> of fiber, there is plenty of light. Behind the scanhead, no light arrives.
>> We work with an inverted IX 81 microscope.
>>
>> Unfortunately the service situation in Brazil is very poor, so if there
>> were any suggestions how to proceed, I would rather try to check the
>> alignment myself.
>>
>> Anyone having experience with the FluoView1000 scanhead?
>>
>> Thank you, Jens
>>
>> Dr. Jens Rietdorf, visiting scientist @ center for technological
>> development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de Janeiro
>> Brasil.
>>