http://confocal-microscopy-list.275.s1.nabble.com/laser-alignment-issue-FluoView1000-tp7584981p7585266.html
ch3. A good guess seems to be 'Spectral Detectors out of tune' which was
Dr. Jens Rietdorf, visiting scientist @ center for technological
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Jens,
> Have you tried redirecting the blue and green channels to the detector
> that is usually “red”?
> after ~28,000 hrs on our FV-1000, the “green” channel started dropping
> out, sometimes entirely, other times with dark bands in the scanned frame.
> The problem persisted when routed to other detectors. The culprit was the
> AOTF.
>
> Good luck,
> Glen
> Glen MacDonald
> Digital Microscopy Center
> Box 357923
> University of Washington
> Seattle, WA 98195-7923 USA
> (206) 616-4156
>
[hidden email]
>
>
>
>
>
>
>
>
> > On Jun 1, 2016, at 10:35 AM, jens rietdorf <
[hidden email]> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Dear colleagues,
> >
> > thank you all for your help on the issue, it turned out a shutter inside
> > the scanhead was blocked. This problem was relatively rapidly addressed
> by
> > the service of Olympus Brasil and Olympus Americas. Very helpful input
> for
> > the differential diagnosis came from Farid Jalali of Olympus Canada.
> >
> > A new (maybe related) issue occurred now, we experience very low
> > sensitivity on detector channels 1&2,
> > channel 3 operates normally. the effect appears to be stronger in the
> > blue/green part of the spectrum than for the red.
> > The FV diagnostic tool does not indicate any abnormalities.
> >
> > Again, all ideas and suggestions are very welcome.
> >
> > Kind regards, Jens
> >
> > Dr. Jens Rietdorf, visiting scientist @ center for technological
> > development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de
> Janeiro
> > Brasil.
> >
> > On Mon, Apr 4, 2016 at 6:39 PM, jens rietdorf <
[hidden email]>
> wrote:
> >
> >> Listers,
> >>
> >> following the installation of a new air condition, we are not succeeding
> >> to get any light after the scanhead of our Olympus FluoView1000 CLSM.
> End
> >> of fiber, there is plenty of light. Behind the scanhead, no light
> arrives.
> >> We work with an inverted IX 81 microscope.
> >>
> >> Unfortunately the service situation in Brazil is very poor, so if there
> >> were any suggestions how to proceed, I would rather try to check the
> >> alignment myself.
> >>
> >> Anyone having experience with the FluoView1000 scanhead?
> >>
> >> Thank you, Jens
> >>
> >> Dr. Jens Rietdorf, visiting scientist @ center for technological
> >> development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de
> Janeiro
> >> Brasil.
> >>
>