Thank you for your reply.

On 17/06/2016 7:10 PM, Michael Giacomelli wrote:
>> I also heard off-list from Q-Imaging, who suggested their R1 (1.4MP CCD,
> cooled).
>
> For what it's worth, the read noise on that camera is worse than many of
> the inexpensive machine vision cameras Edmunds or Thorlabs sell.  Unless
> you must have cooling, it may not be a great choice.
The R1 spec sheet (
www.qimaging.com/products/datasheets/Retiga-R1-Datasheet.pdf ) claims 7
electrons for the read noise.
I have not been able to find the read noise specification for the
machine vision cameras.
Do you have that information, or is there a different parameter from
which RMS read noise can be derived?

>> Is anybody using DSLR cameras for fluorescence imaging?  I know they are
> color and probably less sensitive, but I still wonder...
>
> I've seen these used a lot in medical labs for taking pictures of slides.
> The main problem with a dSLR (besides the software interface) is that the
> sensor form factor is usually very large.  For example, that R1 you linked
> above uses a Sony CCD sensor with an 11 mm diagonal, whereas a typical dSLR
> sensor format is 25-30 mm diagonal.  To maintain the same field of view you
> would either have to change the tube lens or get a ~3:1 expansion
> telescope.  For comparable performance it usually makes more sense to just
> buy a camera with the right sensor size.
>
I was thinking  of dslr cameras with aps-c sensors (which cost much less
than full frame dslr),
but even then you are correct that a zoom will be needed to  fill the
sensor.
However, such adapters are available for under $1000, and the cameras
themselves are under $500.
Nikon makes  2.5x F-mount adapter for their DS-Qi2 (which uses a full
frame sensor) which costs about 450 Euro.
So if the dslr has sufficient  sensitivity, it can be worth it.

Best regards
--aryeh

>
> On Fri, Jun 17, 2016 at 1:07 AM, Aryeh Weiss <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Many thanks to the many people who already responded to my inquiry.
>>
>> On 16/06/2016 11:51 PM, Armstrong, Brian wrote:
>>
>>> Hi Aryeh, I think that you can get a good scientific camera for that much
>>> from Q Imaging. Thor Labs also has CCD cameras around this price point.
>>> I inherited 4 astronomy cameras and was very disappointed with them and
>>> ended up just giving them away
>>>
>> I also heard off-list from Q-Imaging, who suggested their R1 (1.4MP CCD,
>> cooled). I suspect that in Israel it will too expensive, but we are
>> checking that.
>> Thanks for the comment on the astronomy cameras.
>>
>> On 17/06/2016 12:35 AM, Pedro Almada wrote:
>>
>>> Hi Aryeh, have you had a look at the Hamamatsu flash 2.8?
>>>
>>> This is an attractive option, but I am pretty sure it will be too
>> expensive (we will check). I know that the Flash 4.o costs around $9k in
>> Israel. What does it cost in your area?
>>
>> On 17/06/2016 12:56 AM, Michael Giacomelli wrote:
>>
>>> Point Grey periodically updates their evaluation of cost-effective
>>> sensors:
>>>
>>> https://www.ptgrey.com/camera-sensor-review
>>>
>>> The newly added IMX264/IMX265 series sensors obtain a read noise of about
>>> 2.3 electrons and an absolute sensitivity of about 4 photons.  sCMOS will
>>> do better (read noise of about 1 electron on the newest models).  From the
>>> brightness of your specimen you can get a rough idea how well that will
>>> work for you.
>>>
>> I have a Pt Grey Chamelian, but I found that I cannot easily use it with
>> Micromanager. It claims to run with IIDC (DC1394), and it does, sort of.
>> However, at least in my hands, it was difficult, and not all of the modes
>> worked in micromanager (although it worked ok with coriander , which is a
>> program used for testing such things).
>>
>> On 17/06/2016 2:01 AM, Ochoa, Lorenzo F. wrote:
>>
>>> Hello Aryeh,
>>>
>>> I've had good results from the Thorlabs High Sensitivity USB 3.0 CMOS
>>> Cameras. They are relatively cheap ($1.5K) and are compatible with
>>> Micro-Manager:
>>> https://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=4024
>>>
>> This looks interesting. The price is right, at the cost of being uncooled
>> (but cooling for exposures under a second may not be critical). It is
>> supposed to work with micromanager.
>>
>> Is anybody using DSLR cameras for fluorescence imaging?  I know they are
>> color and probably less sensitive, but I still wonder...
>>
>> Again thanks to the many people who replied.
>>
>> --aryeh
>>
>>
>>
>>
>>
>>
>> On 16/06/2016 10:26 PM, Aryeh Weiss wrote:
>>
>>> I was asked if I can recommend a camera for fluorescence microscopy for
>>> under $4k.
>>> It is not single molecule, but if they can see it, they would like the
>>> camera to see it.
>>>
>>> I found a similar question on the list from 2014
>>>
>>> https://lists.umn.edu/cgi-bin/wa?A2=ind1403&L=confocalmicroscopy&P=R18770&1=confocalmicroscopy&9=A&J=on&d=No+Match%3BMatch%3BMatches&z=4
>>>
>>> but this is a moving target, and my price constraint is higher, so I am
>>> wondering what else is out there.
>>>
>>> Are there any sCMOS  cameras that qualify? Does anyone have experience
>>> with the less known brands (eg cameras geared toward amateur astronomy)?
>>> I have a strong preference for cameras that will work with micro-manager.
>>> I would appreciate finding out what solutions people on the list have
>>> found.
>>>
>>> Thanks in advance
>>> --aryeh
>>>
>>>
>> --