Re: Shearing Interferometer with Ti-Sapph
Posted by
jcv2@uw.edu on
Sep 21, 2016; 4:28am
URL: http://confocal-microscopy-list.275.s1.nabble.com/Shearing-Interferometer-with-Ti-Sapph-tp7585684p7585685.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on
http://www.imgur.com and include the link in your posting.
*****
Here's a shot in the dark here for you, since I have never tried doing what you asked about.
Short pulse lasers only interfere over a short distance called their coherence length. It might only be 10-30 microns or so depending on the laser's bandwidth. If the path length difference of the two interfering beams is more than this 10-30 microns, then no fringes.
If the interferometer works by beam 1 reflecting off the front of the glass wedge (going through no glass) and beam 2 reflecting off the back surface of the glass at a little different angle (going through the glass twice), then the optical path length of beam 2 is quite a bit more than that of beam 1. Probably by much more than 10-30 microns since the wedge is probably at least 1 mm thick where you use it. I would guess no fringes.
Basically, I am asking whether your interferometer is designed for short pulse lasers. I bet the Thorlabs folks could weigh in on it if you email them.
Anyway, isn't it easier to check collimation by shooting the beam across the room and checking the diameter is constant?