You could check the collimation with the Ti-S in CW mode (i.e. not mode locked)?
HTH
Mark
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hello, we recently purchased a shearing interferometer from Thorlabs (Part # SI254) to test collimation after beam expansion in our two-photon microscope setup. The strange thing is we cannot seem to get an interference pattern either before or after beam expansion (varying the distance between the lenses in the beam expander). We do see the pattern clearly if we try on a simple laser diode. Does anyone have experience looking at the shearing interferometer pattern with a Ti-Sapphire laser or have any ideas what may be causing our problem?