The way I check collimation in a Ti:Saph is using a cheap CCD or CMOS camera placed at a few different points in the beam. As long as the IR filter has been removed, the camera should be able to 'see' the beam projected onto the array. Do note
you will want some good reflective ND filters on there to keep from burning the camera! You want to take pictures at several points along the path, or just use mirrors to extend the path while leaving the sensor stationary. You can get a rough measurement
of spot size vs. distance which will give you some indication of where the Rayleigh range of the beam is located.
And yes, coherence length of a Ti:Saph is typically ~30um, so it won't work with a shearing interferometer. Also, running the Ti:Saph in CW may produce a slightly different vergence out of the laser. Finally, don't sweat collimation too much, you can always
put a pair of good achromats into the beam as a telescope to adjust the beam waist position.
Craig