Andreas,
Fair enough. When I worked with beads I used 60nm beads to side-step the convolution issue. What bugs me is that people seem to accept rather
poor resolution figures as being ‘acceptable’ when they are not. We have demonstrated 255nm FWHM resolution on collagen in SHG at 830nm (Guy Cox & Colin Sheppard, 2004. Practical limits of resolution in confocal and non-linear microscopy. Microscopy Research
& Technique, 63, 18-22). Theoretical resolution in these conditions was 210nm, and frankly I’m disappointed that we didn’t attain it. 2-photon should be comparable (actually better since it’s incoherent). These results should be attainable in everyday
use with a modicum of care.
Guy
Guy Cox, Honorary Associate Professor
School of Medical Sciences
Australian Centre for Microscopy and Microanalysis,
Madsen, F09, University of Sydney, NSW 2006
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Andreas Bruckbauer
Sent: Friday, 30 September 2016 7:02 AM
To: [hidden email]
Subject: Re: Two Photon Microscope Low Signal Troubleshooting
***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Guy,
As I imaged at 800 and this should be treated as high NA, one would use 0.325 * lambda / (sqrt(2) * NA^0.91 and multiply with 2*sqrt(ln2) to convert from 1/e radius to fwhm (Zipfel et al 2003) which then gives 293 nm. So l calculated my own PSFs taking the
vector properties and polarisation into account and got 320 nm. Convolution with the bead size of 175 nm gives roughly 365 nm (assuming Gaussian profiles). The difference between theory (365 nm) and real life (410 nm)was only 12%. It did not matter as the
cells were 5 um diameter and this was all we wanted to image.
Best wishes
Andreas
From:
[hidden email]
Sent:
29/09/2016 18:04
To: [hidden email]
Subject:
Re: Two Photon Microscope Low Signal Troubleshooting
***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
In 2-photon microscopy your resolution (FWHM) should be 0.5 * wavelength / (1.414 * NA) . At 900nm and NA 1.05 this works out at 303nm. Surely if you are
just imaging beads you should be able to do better than 410nm?
Guy
Guy Cox, Honorary Associate Professor
School of Medical Sciences
Australian Centre for Microscopy and Microanalysis,
Madsen, F09, University of Sydney, NSW 2006
From: Confocal Microscopy List [[hidden email]]
On Behalf Of Andreas Bruckbauer
Sent: Thursday, 29 September 2016 5:18 PM
To: [hidden email]
Subject: Re: Two Photon Microscope Low Signal Troubleshooting
***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
@ Andreas:
For the PS Speck beads on a slide, when you say that the image is not too far off the expected values do you mean the expected spatial resolution values of the system?
Yes, I get about 410 nm fwhm in xy and about 1.2 um in z with a 25x 1.05 NA objective when the back aperture is nicely filled. If the z resolution is bad, you might not see small objects but will do just fine imaging larger ones. Also check if you are bleaching
the sample before you actually see it, but in this case I would expect a short spike in fluorescence.
Best wishes
Andreas
From:
[hidden email]
Sent:
29/09/2016 07:25
To: [hidden email]
Subject:
Re: Two Photon Microscope Low Signal Troubleshooting
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Thank you all for your responses.
@ Michael:
Our Ti-Sapphire has 100 fs pulses and is tuned to 900 nm. We use only protected silver mirrors and the following lenses from Thorlabs:
Initial Beam expansion with: