***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****Hi George,As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that?Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining.Thanks!Sripad RamOn Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****Hi Claire,
Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.
I suspect the TSA reaction runs to completion usually, with one of these running out:
* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)
* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)
* fluorescent tyramide (least likely?)
* H2O2 (also unlikely?)
I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).
Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80- conjugate-65r-s118.html
Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.
I'm a big fan of TSA ever since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).
One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).
Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.
Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).
One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish http://biocare.net/product/peroxabolish/
https://www.ncbi.nlm.nih.gov/pubmed/21929847
Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 16. Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue using laser scanning cytometry. Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker S, Burke GW, Tzakis AG, Ichii H. Author information: (1)Miami Transplant Institute, University of Miami Leonard M. Miller School of Medicine, Miami, FL, USA. There is abundant evidence that immune cells infiltrating into a transplanted organ play a critical role for destructive inflammatory or regulatory immune reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune cells remains challenging due to a lack of objective methodology. Laser scanning cytometry (LSC) is an imaging-based methodology that performs quantitative measurements on fluorescently and/ or chromatically stained tissue or cellular specimens at a single-cell level. In this study, we have developed a novel objective method for analysis of immune cells, including Foxp3(+) T regulatory cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy sections using iCys® Research Imaging Cytometer. The development of multiple immunofluorescent staining was established using FFPE human tonsil sample. The CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using iCys and compared with the results from conventional flow cytometry analysis (FCM). Our multiple immunofluorescent staining techniques allow obtaining clear staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with those obtained by FCM. This method was also applicable for liver, small intestine, kidney, pancreas, and heart transplant biopsy sections and provide an objective quantification of Tregs within the grafts. DOI: 10.3727/096368911X586747 PMCID: PMC3777543 PMID: 21929847 [PubMed - indexed for MEDLINE]and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).
I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.
Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):
Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/ ) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.
I see in SuperBoost PDF only 7 colors (7 Alexa's)
https://tools.thermofisher.com/content/sfs/manuals/ tyramide_superboost_kits_man.
to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full
Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).
Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.
enjoy,
George
On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?
https://www.thermofisher.com/
order/catalog/product/B40915
I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?
I would love to hear what people think.
Sincerely,
Claire
-- George McNamara, PhD Houston, TX 77054 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75/ http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/ 44962650
| Free forum by Nabble | Edit this page |