***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****Fluorescent light cannot produce interference. So this has to be reflection. It could be due to something with filters or from reflection from something inside. Have you try to remove the sample? Defocus the sample? Try single line scanning?
From: Aryeh Weiss <[hidden email]> on behalf of Aryeh Weiss <[hidden email]>
Sent: Sunday, November 20, 2016 9:49 AM
To: MODEL, MICHAEL; [hidden email]
Subject: Re: interference fringes in SP5 scans at 10xThank you for your reply. This is fluorescence, not reflectance, imaging. The passbands are set to reject the excitation wavelengths.
--aryeh
On 20/11/2016 16:18, MODEL, MICHAEL wrote:
Hello Aryeh,
You are doing fluorescence or reflection? If reflection then the fringes are almost unavoidable. For example, light reflecting from the front and from the back of the coverslip should interfere because of the long coherence length of the laser
Mike
From: Confocal Microscopy List [hidden email] on behalf of Aryeh Weiss [hidden email]
Sent: Sunday, November 20, 2016 7:25 AM
To: [hidden email]
Subject: interference fringes in SP5 scans at 10x***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...
[hidden email]: listserv archives. confocalmicroscopy
We are observing what appear to be interference fringes form reflected light in our Leica SP5, with our 10x objective.
Example images can be seen using this link:
There is a README file that says which excitation lines were used.
https://drive.google.com/drive/folders/0Bytu294eIohjVHQ1elVT SElscjg?usp=sharing
Basically, when the 405nm excitation is used alone, we do not see the fringes. When the other lines (ie the "visible" lines, as opposed to what Leica calls UV, which is really V) are used, there are fringes even in the channel which is set to pass wavelengths shorter than the shortest excitation in use (in our case 488nm).
So, in brief, the "visible" laser lines all produce interference fringes in all channels. The fringe spacing changes with excitation wavelength (as expected), but not with scan speed. They are not apparent when working with 40x oil. However, I note that with the oil objective, there is much less reflection from the coverslip . We do not have other air objectives with which to test the system.
The fringes appear with various samples.
It is quite possible that this has always been there, but no one noticed because confocal at 10x is not done very often.
If anyone on the list can tell us more about these fringes, we would appreciate it.
Thanks in advance.
--aryeh
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051
| Free forum by Nabble | Edit this page |