Răspuns: AO and PI

Posted by Ozana Maria Petraru on
URL: http://confocal-microscopy-list.275.s1.nabble.com/no-subject-tp7586114p7586121.html

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Acridine orange is an intercalating dye that can permeate both live and dead cells. AO will stain all cells to generate green fluorescence. Propidium iodide can only enter dead cells with poor membrane intergrity => all live cells fluoresce green and all dead cells fluoresce red. I can not use DAPI because I do not have a UV filter.
 I ve used I3 filter (blue) to to excite both fluorochromes and also i ve tried the mixture on Bacillus subtilis (it did not work).  I attached some pictures: S. aureus AO+PI (when the mixture work), S. aureus AO-PI DIC 1 and S. aureus AO-PI FLUO 1 (when the mixture did not work) - photos pair. I used DIC and FLUO to the same microscopic field to highlight that there are red cells and some green cells with a poor fluorescence and some cells with any fluorescence at all.
Thank you!
 Ozana Petraru, Iasi, Romania
 Imgur: The most awesome 



images on the Internet
















În Joi, 8 Decembrie 2016 2:23:36, "MODEL, MICHAEL" <[hidden email]> a scris:


***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Ozana,

What exactly stopped working?

Try killing all bacteria with heat, check that PI works. DAPI for example, doesn't stain every cell type, and I don't know about AO. If all you bacteria are alive and AO for some reason stopped working you might not see anything.

What else:
Make sure there are not too many bacteria so the dye doesn't get depleted (that's unlikely though).
Check your microscope settings.
Try other cells.
Post pictures, that might help.

Good luck!



From: Confocal Microscopy List <[hidden email]> on behalf of Ozana Maria Petraru <[hidden email]>
Sent: Wednesday, December 7, 2016 5:12 PM
To: [hidden email]
Subject:
 
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy

Heloo!
            I am performing a membrane integrity test on Saphylococcus aureus using a mixture of two fluorochromes: propidium iodide and acridine orange. The cells grow in a MHB medium, washed and resuspended in PBS pH-7,2.
Acridine orange and propidium iodide were added to 100 µl aliquots of these cells, followed by incubation in the dark for 15 minutes.
           In the past this method worked, so the cells with damaged membrane were stained in red and the cells with intact membrane in green. In the past two  months this double staining  does not  work anymore, and I do not know why?!!!!  The  cells are no longer staind in green. In the best  case  the cells  have a low fluorescence.
 I ve rebuilt stock solutions but nothing improved. 
 It can be the PBS?
 Do you have any idea how to solve or  find the problem?
Thank you!
 
 
Ozana Petraru - Master's degree  Microbial and cell biotechnology
 Faculty of Biology
"Al. I. Cuza" of  Iasi University