Hi Julia,
I used to have these lines block with condensed water. This is easy to diagnose, if you disconnect the blue line at the CO2 mixer, water will often come out. Some compressed air gets rid of the rest.
Elevating the bottle to a shelf above the stage seems to have prevented this from happening further.
Doug
Sent from my iPhone
> On Apr 3, 2017, at 5:18 AM, Julia Edgar <
[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear All
> I’m using a Zeiss spinning disc confocal microscope with small heated incubation stage and ‘gassing’ lid.
> The problem is that the pH of the media changes over time and becomes rather alkaline (turns blue pink). Motor proteins don’t behave normally when pH is not physiological, so this is potentially a big problem.
>
> Do you have any tips to ensure the 5% CO2 reaches the imaging dish (a 35 mm Petri-dish) that sits between the incubation insert and the gassing lid?
> The tubing connecting the gassing lid to the water bottle (into which the CO2 is pumped) is rather long. Would it help to shorten it?
> Does the volume of water in the water bottle matter?
> How can I ensure the little inlet pipes in the gassing lid are not blocked?
>
> Many thanks
> Julia
> University of Glasgow.