Hi Lu,
We have means of removing laser speckle within fiber, up to 600 um core.
Optical power loss is greatly reduced compared to diffusers and less
complicated than getting out of the fiber.
Typical integration time with >1/5 spatial noise is in the 5 ms range.
Using a non circular fiber, it is possible to also produce a flat top ,
although laser coupling options in standard fiber may be less expensive.
Eric Girard
-----Message d'origine-----
De : Confocal Microscopy List [mailto:
[hidden email]] De
la part de Yan, Lu
Envoyé : August-14-17 11:24 PM
À :
[hidden email]
Objet : Re: Wide field, Epi illumination, overfilling the aperture?
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Agreed, Mike.
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On
Behalf Of MODEL, MICHAEL
Sent: Friday, August 04, 2017 4:31 AM
To:
[hidden email]
Subject: Re: Wide field, Epi illumination, overfilling the aperture?
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Dear Lu,
In practice, we focus the collector lens at whatever plane gives us the best
combination of uniformity and brightness. But it is never perfectly flat
across the field unless, as George McNamara said in his post, you use a
scrambler or something. I suppose that by simply projecting something as
variable as mercury arc you cannot expect to get perfect illumination.
Mike
________________________________
From: Confocal Microscopy List <
[hidden email]> on behalf
of Yan, Lu <
[hidden email]>
Sent: Friday, August 4, 2017 12:45 AM
To:
[hidden email]
Subject: Wide field, Epi illumination, overfilling the aperture?
*****
To join, leave or search the confocal microscopy listserv, go to:
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Cmmodel%40KENT.EDU%
7Ca9ba5fb033a34939f9d708d4daf3d3ad%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sd
ata=wbx4UwbUCyfd7kwJdrd5n%2FnGXlDKROgBspzdg4o0igM%3D&reserved=0
Post images on
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Cmmodel%40KENT.EDU%7Ca9ba5fb033a34939f9d708d4daf3d3ad%7Ce5a0
6f4a1ec44d018f73e7dd15f26134%7C1&sdata=B4%2BxjBLa9gf30TE7Uy4np07eadfth81QXu0
vG9qbOqk%3D&reserved=0 and include the link in your posting.
*****
Hi,
I was searching for back aperture (or exit aperture plane) and the back
focal plane of high NA objective lenses, and found this thread that Kyle
posted in end of 2015:
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fconfocal-microscopy-list.588098.n2.nabble.com%2FObjective-back-focal-plane-vs-exit-pupil
-plane-locations-td7584559.html%23a7584565&data=01%7C01%7Cmmodel%40KENT.EDU%
7Ca9ba5fb033a34939f9d708d4daf3d3ad%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sd
ata=N9twDfsIfwM4sXG6QVDH2%2FbyDzlxxjN%2BzQgEcrr4trA%3D&reserved=0, in which
all agreed most of cases, these two planes are not coplanar.
I understood that, for confocal scope, one would want to 'overfill' the back
aperture of an objective, as described in several textbooks, to utilize the
full NA of the lenes, and achieve smallest spot size possible. There, since
the laser beam comes in collimated, it of course also overfills the back
focal plane which often lays inside of the objective lens (I found this
website providing Zemax models of several Nikon lenses based on publicly
released documentations:
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Ffigshare.com%2Farticles%2FZemax_optical_design_files_of_microscope_objectives_tube_lens
es_and_Fourier_imaging_setups%2F1481270&data=01%7C01%7Cmmodel%40KENT.EDU%7Ca
9ba5fb033a34939f9d708d4daf3d3ad%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1&sdata
=cGMeYj3U8Twi0kiG2dOcGXeZN1TVADipZ%2F1zZKm33QA%3D&reserved=0 ). I have
simulated several of them, and found that the 'back focal plane' is indeed
residing inside of the lens. Some of them are actually almost in midway of
the lens housing.
Based on these information, I was wondering, for wide field scope,
epi-illumination (non-critical illumination), what would be the best
practice to choose the relay lenses to focus onto this back focal plane to
get most uniform, and possibly large illuminated FOV, considering there is a
hard stop up stream, i.e. the exit aperture of the lens. It seems to me that
depending on the distance between these two plane, and the size of the exit
aperture, there is a limited NA for lens that focuses the collimated laser
beam onto the back focal plane. Is there an 'overfilling'-equivalent concept
for epi-illumination? Does it matter if one comes with quite large cone
angle to focus onto the back focal plane, and the beam is clipped by the
back aperture of the objective lens? Would the amount of clipping affect the
illumination quality in general? I guess this kind of clipping will play
more important role in TIRF or SIM case where one wants to focus beam(s)
onto the edge of the back focal plane of the objective. Smaller cone angel
will give you small beam size (collimated) on the sample plane), but large
cone angle will be clipped by the back aperture of the objective. I am
trying to collect all relevant thoughts from experts here on the list, and
any inputs will be welcome. I have to get several concept clear in my head.
Thanks,
Lu
---