Dear Pete,
I was wrong, we actually have a single mode and not a multi-mode fiber. May
I ask what is to be expected in terms of coupling efficiency for single
mode fibers?
Thank you for your input,
All the best,
Emmanuel
On 6 September 2017 at 22:15, Peter Brunt <
[hidden email]>
wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> *commercial (non-commercial) post*
>
> Hi Emmanuel
> Are you running your lasers at full power? What combiner are you using?
> I can't comment so much on the power you have at your objective but on your
> combiner side the values look low. Going into a multi-mode fibre should be
> like a barn door for your lasers so if we imagine a ~5% (x2) loss due to
> combining optics and an ~80% coupling efficiency, you should still be
> getting a ~72% average coupling efficiency into the fibre. It sounds to me
> like the setup may need to be realigned. If you still don't see an
> improvement then one other possibility is that the 405nm laser has degraded
> your fibre due to solarisation and the fibre needs replacing.
>
> I hope this helps.
>
>
> Kind Regards
>
> Pete Brunt
>