a room within the fish facility. Now the fish facility requires 26 degC for
the fish. Will it (26 deg C and humidity < 60% and both maintained quite
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Right, with optics a stable temperature is more important than an absolute
> temperature, as long as the temperature is around room temp. For humidity,
> the biggest concern is condensation damaging the optics. Basically, the
> concern is that say the cooling had a hard time and the room is at 26oC and
> 70% humidity. If the temperature then drops to 20oC during an AC cycle,
> the relative humidity will then be approaching 100% inside the scope, which
> will definitely damage antireflective coatings and dichroic mirrors. That
> is why spec sheets say 60% humidity max, as this give plenty of safety
> buffer for temperature fluctuations without risking condensation. A data
> logger is invaluable, both for your own records, and any insurance claims
> if necessary. Also, past 60% humidity you will also notice that near-IR
> mode-locked lasers don't perform as well.
>
> -Ben Smith
>
> On Thu, Oct 5, 2017 at 10:15 PM, Balakrishnan Kannan <
>
[hidden email]
> > wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Dear Benjamin,
> >
> > Thanks. Being in a tropical climate we have a high ambient humidity,
> > typically >70%. Some of the lab areas have the AHU that can maintain the
> > humidity around 55-60%. If I ask for lower than that it requires portable
> > dehumidifier. It adds to the power bill. So we want to be optimal so as
> to
> > maintain the optics free of mould and at the same time reducing
> > additional costs.
> >
> > Temperature can be maintained in the range of 22-23 deg C. I believe that
> > is fine as long as it does not vary much.
> >
> > Best regards,
> > Core Imaging Facility Staff
> > NTU
> >
> >
> > On 6 October 2017 at 12:07, Benjamin E Smith <
>
[hidden email]>
> > wrote:
> >
> > > *****
> > > To join, leave or search the confocal microscopy listserv, go to:
> > >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > > Post images on
http://www.imgur.com and include the link in your
> > posting.
> > > *****
> > >
> > > Shoot for 40% relative humidity and 21.5oC. Humidity can normally be
> 20%
> > > to 60% while the temperature you will want to be as stable as possible
> to
> > > reduce thermal creep in the optics. Most microscopes have a spec sheet
> > > that states the range they can optimally operate in.
> > >
> > > On Thu, Oct 5, 2017 at 8:18 PM, Balakrishnan Kannan <
> > >
[hidden email]>
> > > wrote:
> > >
> > > > *****
> > > > To join, leave or search the confocal microscopy listserv, go to:
> > > >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > > > Post images on
http://www.imgur.com and include the link in your
> > > posting.
> > > > *****
> > > >
> > > > Dear Microscopists,
> > > >
> > > > Is there a standard temperature and humidity setting for an imaging
> > core
> > > > suite where confocal and wide-field microscopes are stationed?
> > > >
> > > > Thanks for your time.
> > > >
> > > > Best regards,
> > > > Core Imaging Facility Staff
> > > > NTU
> > > >
> > >
> > >
> > >
> > > --
> > > Benjamin E. Smith, Ph. D.
> > > Imaging Specialist, Vision Science
> > > University of California, Berkeley
> > > 195 Life Sciences Addition
> > > Berkeley, CA 94720-3200
> > > Tel (510) 642-9712
> > > Fax (510) 643-6791
> > > e-mail:
[hidden email]
> > >
http://vision.berkeley.edu/?page_id=5635 <
http://vision.berkeley.edu/>
> > >
> >
>
>
>
> --
> Benjamin E. Smith, Ph. D.
> Imaging Specialist, Vision Science
> University of California, Berkeley
> 195 Life Sciences Addition
> Berkeley, CA 94720-3200
> Tel (510) 642-9712
> Fax (510) 643-6791
> e-mail:
[hidden email]
>
http://vision.berkeley.edu/?page_id=5635 <
http://vision.berkeley.edu/>
>