You say you have had the system for more than a year, but was the system delivered and installed with these problems? If yes, unless someone in your lab has tinkered with the system and played with the compents without Andor's supervision or permission, then Andor (as would any other manufacturer/vendor) has a responsibility to make this work even if it means replacing large components. A new system should be installed working at or better than specification. I don;t know the law in Germany, but if this were in the US, depending on what state you were in and/or what the PO says, this may be a contractual requirement.
Michael Cammer, Research Scientist, DART Microscopy Laboratory
NYU Langone Health, 540 First Avenue, SK2 Microscopy Suite, New York, NY 10016
C: 914-309-3270
[hidden email]<mailto:
[hidden email]>
http://microscopynotes.com/https://med.nyu.edu/research/research-resources/scientific-cores-shared-resources/microscopy-laboratory________________________________
From: Confocal Microscopy List <
[hidden email]> on behalf of Anje Sporbert <
[hidden email]>
Sent: Tuesday, December 5, 2017 11:55 AM
To:
[hidden email]
Subject: Re: Best dual camera CSU-W1 Spinning-Disc solution?
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=V9JR5MP8LvSymZyiwzQJQylolbSEYZRG4pl_dMfYWyQ&s=FPDAFU0yddjOH1L1vfhU-0_taBYd0OKXyKjC_LEumdM&e=Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=V9JR5MP8LvSymZyiwzQJQylolbSEYZRG4pl_dMfYWyQ&s=ANXhGUlJAbyZU8YWvbbcAshsaVWrfkaO6NVEtuvGA5E&e= and include the link in your posting.
*****
Dear Kai, dear all
we have a CSU-W1 spinning disk setup provided by Nikon (microscope, autofocus and software) and Andor (EMCCD camera and CSU-W1 unit with Borealis) for more then one year.
All in all we are satisfied with the performance of the system and the flexibility the Nikon NIS software offers.
However, there are a few technical problems with the CSU-W1 which Andor could not manage to solve until now although we had several visits from engineers and they tried hard.
First, somehow there seems to be some dust inside the SD unit. Although the system has now cleaned and additionally sealed by the engineers, we see after some time dust particles again on the image.
Second, there is quite some lateral shift between two images if we change the dichroics between image acquisition (one camera). This shift is larger then should be expected - according to Andor - and it is larger if we use the 20x objective and smaller if we acquired the same image with an 100x objective.
Third, we see some kind of weak ghost images: parts of the image structures appear again in other parts of the image but as much weaker structures. We see this with all lasers and filter combinations. The ghost images move in opposite directions if we move the sample. There seem to be some internal reflections the source of this but until now there is no solution to this problem.
It would be very usefull to hear if other places have similar problems with the CSU-W1 unit or if these are problems only with the Andor/Borealis combination? Are there solutions for these problems around?
Best wishes, Anje
------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================