Hi Eric,
yes, new Zeiss objectives are M27 x 0.75, but the thread adapter should have
a flange so it can be tightened in the objective turret. See the following
pictures (ask your Zeiss rep, thy might lend you one):


https://www.micro-shop.zeiss.com/pimages/000000-1095-168_lg.jpg

https://www.hofstragroup.com/media/product_images/productimage-picture-zeiss
-intermediate-ring-rms-w-0-8-to-m27-objective-adapter-2392.jpg
Best, zdenek

--
Zdenek Svindrych, Ph.D.
Research Associate - Imaging Specialist
Department of Biochemistry and Cell Biology
Geisel School of Medicine at Dartmouth


---------- Původní e-mail ----------
Od: Eric Marino <[hidden email]>
Komu: [hidden email]
Datum: 3. 2. 2018 9:12:30
Předmět: Zeiss LSM880
"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Good Morning All,



Is the LSM880 objective nosepiece thread size an M27 x 0.75? I’m trying to
use an Olympus objective with a thread size of W20.32×0.706 (RMS). And I’m
considering the RMSA3 - Adapter with External M27 x 0.75 Threads and
Internal RMS Threads from Thor Labs. https://www.thorlabs.com/thorproduct.
cfm?partnumber=RMSA3



Any suggestions would be greatly appreciated.