Re: Readout noise in sCMOS cameras

Posted by Jeremy Adler-5 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Readout-noise-in-sCMOS-cameras-tp7588012p7588018.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****



Dear James,
a minor point from your interesting post.

"Poisson Noise get bigger as the signal increases!"

This is true as an absolute measure of the size of the variation around the mean but your statement could be misinterpreted as a preference for detecting fewer PEs.
Perhaps adding -      but is smaller as a fraction of the signal
An average of 16 PEs has a standard deviation of 25% while for  256 PEs the SD drops to around 6% and to 3% for 1024.


Jeremy Adler
IGP, Uppsala U, Sweden
====================================




+46 70 1679349

http://www.biovis.uu.se







-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of JAMES B PAWLEY
Sent: den 6 mars 2018 07:22
To: [hidden email]
Subject: Re: Readout noise in sCMOS cameras

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Sripad,

In a CMOS chip, every pixel has its own read amp. All of these vary slightly in gain and DC-offset. So the raw output from a a black (no light) image would have a noise term related to how much the offsets of the pixel amps varied and a uniform white image would have Poisson noise on the photo charge plus a noise term mostly related to the variation in the gains of the pixel amplifiers.
In an sCMOS chip to these uncertainties must be added variations in the gain and offer of the 4,000-plus separate ADCs mounted at the edges of the chip. An effort is made to correct for the multi-amplifier and multi-digitizer noise by “flat fielding” the raw data from the chip using data from previous “black” and “white” images, The system works quite well but as the white image always involves a lot of charge, its Poisson noise (sqrt of n) is large and this can skew the results. So can using the chip at a different temperature, dwell time or pixel clock than was used for the “black” and “white” images. Other sources on non-“Gaussian”  noise include “hot pixels” (perhaps leaky photodiodes that are sometimes flagged and removed by the camera system software).

Indeed, the noise spectrum in these low-light systems is almost never “Gaussian”. Even if the electronic noise (that signal variation which becomes evident when reading the same pixel with no light signal) seems Gaussian, it is usually caused by Poisson Noise (Or Johnson noise) affecting the small number of electrons that constitute the (fairly table currents passing through the elements of the charge amplifier. And of course, at signal levels of more than a few dozen photoelectrons, Poisson Noise on the PE number soon dominates most other noise sources (not hot pixels).

Gaussian noise is just easier to think about, and easier to model. We should remember that in low-light photodetectors, it is almost never appropriate. (Poisson Noise get bigger as the signal increases!).

Best,,

JP
James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone 1-604-885-0840, cell 1-604-989-6146



On Mar 5, 2018, at 9:54 AM, S Ram <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello Gerhard,
This is a slightly off-topic question in connection to your recent response to the thread on the choice of sCMOS cameras.

You made a comment that the distribution of noise in sCMOS is not Gaussian.

Can you clarify whether you meant noise during the readout process (charge to voltage conversion step)? If it is not Gaussian, what is the underlying noise process? Is there any literature that you can point me to?

Thanks.

Sripad