Hi Gary,

The R.I. will be the issue more than the FEP tubing holding up against BABB.  If you need something to hold your BABB cleared sample then you should be looking for some kind of boro-silicate glass tubing as this will match up with the BABB a lot better (still not exact but better than FEP would).  I know Zeiss were talking up boro-silicate for use on their Z.1 as a way of sealing things like 3DISCO away from the sample chamber/objective.  I did track some down to a company in Germany who would sell various diameters and custom size the length for you.

Cheers
Mark

_______________________________________
 
Mark Scott
Senior Microscopist
Institute of Molecular Bioscience
Level 6 North, 306 Carmody Road (Building 80)
The University of Queensland
Brisbane, Queensland, 4072, Australia
[hidden email]
 
 


On 26/4/18, 12:01 am, "Confocal Microscopy List on behalf of Gary Laevsky" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi All,
   
    Does anyone know if FEP can stand up to BABB?  The obvious test is to put
    some tubing in BABB, but the corrosion could be slow enough that I can't
    see it, and I don't want to contaminate the objective (lightsheet
    application (and I do know this will be an optical nightmare)).
   
    Thanks.
   
    --