Tissue imaging auto fluorescence and bright dot-like signal

Posted by Lu Yan on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Tissue-imaging-auto-fluorescence-and-bright-dot-like-signal-tp7588326.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear listers,

I have been imaging mouse brain tissue samples for the last a few days using multiple wavelengths laser as illumination (from 405 to 750 nm). Initially, the targeted genes in the brain slice were labeled with different dyes, e.g. Cy3, Cy5, and Cy7 etc., and I observed quite bright dots in corresponding channels, and thought they were RNA molecules. However, further examination showed that those dots showed up even in 'plain tissue' samples (no probes at all), and they virtually showed up in almost all fluorescence channels.

The intensity was quite high that we could not distinguish the fluorescent probes signals. The plain tissue was snap-frozen and sectioned, and fixed with just methanol on coverglasses. I had tried different part of brain, different mouse, and they all showed similar results, i.e. bright dots cross the tissue sample, sometimes dense, sometimes sparse. They also distributed along the z axis, over about 10 um which is the sample thickness. In some region, I can see larger clusters giving much stronger signal. I also tried mouse liver samples, and the signal level was very similar to those from brain samples.

I was wondering if anyone had similar experience in imaging tissue samples. What could be causing this kind of signal? Any input is welcome and appreciated.

Thanks in advance,
Lu