Hi Rory,
8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz
scanners are usually not that bad even at maximum amplitude because the 8
kHz tone is a lot harder to hear and essentially all the harmonics are out
of audible range.
For 4 KHz at larger deflection angles you either need hearing protection or
to really isolate it from the room. A metal box around the scanner and a
window on the input port so that air cannot pass in/out of the scanner
helps a lot. This is how most commerical systems work.
I don't think the model matters; the noise comes from pushing air, so for a
given mirror size and deflection angle the noise power is going to be
almost constant.
Mike
On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Confocalists,
>
> We’re trying to circumvent the annoyance of resonant mirrors screaming at
> 8 kHz. I was wondering whether anyone has a recommendation for constructing
> soundproof containment?
>
> Generally were using inexpensive resonant mirrors from EOPC:
>
http://www.eopc.com/sc30.html>
> Also, does anyone know whether the more costly resonant scanners used for
> confocal/multi-photon laser scanning microscopies are any better in this
> regard? Were not hugely concerned with other attributes (simply for light
> sheet microscopy stripe suppression) just the piercing noise.
>
> Many thanks in advance!
>
> Rory
>
> …………………………………………………………
>