Re: Tetraspeck beads for PSFs **vendor reply**

Posted by Kilgore, Jason A. on Aug 03, 2018; 9:05pm
URL: http://confocal-microscopy-list.275.s1.nabble.com/Re-Tetraspeck-beads-for-PSFs-vendor-reply-tp7588497.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


** Vendor reply **

Polystyrene microspheres, such as the TetraSpecks, will swell in the presence of certain solvents.  So you'll want to use an aqueous-based mounting media, preferably one that cures (such as ProLong products or Fluoromount-G) instead of  using an organic solvent-based one, like Cytoseal-60.  When dye-labeled microspheres like these swell, the dye is free to diffuse out of the polystyrene matrix.

Cheers,

Jason


Jason A. Kilgore
Technical Application Scientist
Molecular Probes / EVOS Tech Support
Thermo Fisher Scientific
 
1-800-955-6288 then option 4, then option 3, then option 2.
Or dial direct at +1 541 335 0353
[hidden email]
 
This communication is intended solely for the individual/entity to whom it is addressed. It may contain confidential or legally privileged information.  Any unauthorized disclosure or copying is prohibited and may be unlawful. If you have received this communication in error, please notify the sender immediately and delete it from your system.
 
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jean Ross
Sent: Friday, August 03, 2018 12:49 PM
To: [hidden email]
Subject: Tetraspeck beads for PSFs

CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe.


*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=F8IDbXdJNbnAT6ffW_HjXIuMlAG3s16jAZb0eldCiQU&s=_ngqUkuraSCI6zAmWzz6NEeoPtsEIhu0PMk_Stm2Gko&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIBaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=F8IDbXdJNbnAT6ffW_HjXIuMlAG3s16jAZb0eldCiQU&s=O2T49sxgjykKvUtqbrfKaSktZURKhrNNnPd3CQna8Gc&e= and include the link in your posting.
*****

Hi Everyone,

I have been preparing 0.1um Tetraspeck bead slides to use when measuring point spread functions on our Zeiss Elyra PS1 super resolution microscope
and Zeiss LSM880 confocal with Airyscan.   The problem is that over time
the beads appear to swell to about 3 times in size and lose much of their fluorescence.  I have tried many different mounting medias (hardening and
nonhardening) all with the same result.  Has anyone else experienced this problem?  Any suggestions to solve the problem would be appreciated.

Thanks,
Jean

--