fluorescence for long periods (years). What ever is in the mix (silicone +
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Jean,
>
> most polystyrene fluorescent beads are susceptible to different mounting
> media, which can ruin the fluorescence of the beads within anything from
> minutes to days. Depending on the mounting media used, you can observe the
> beads swelling, the dye bleeding out into the media or the fluorescence
> simply getting quenched. This is not really an issue, however, if you
> either mount the beads in distilled water or air. This might cause small
> aberrations (refractive index mismatch, reflections at the coverslip-medium
> interface), but they tend to be minimal. If mounting in water, just prepare
> as usual and use water instead of your typical mounting medium. If using
> air, I typically used a little bit of mounting medium (spread of in a
> ring-like manner around the circumference of the coverslip), to form an air
> bubble on the inside, but still keep them isolated from the surrounding
> environment.
>
> Mounted this way, I have maintained some fluorescent bead samples for
> several years now, with only a minor decrease in fluorescence intensity.
> Just keep them at a stable room temperature and out of any light, and they
> should be fairly stable. Just make sure not to drop them... which was the
> bane of most of my long-lived samples.
>
> In my experience, having a reliable sample, of which you know precisely
> what it looks like on your microscope(s) is more important than having
> absolutely perfect PSFs, but remaking your samples every couple of days or
> weeks.
>
> Good luck!
>
[hidden email]
>
> >>>>>>>>>><<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<
> Dr. Nicolai T. Urban
> Max Planck Florida Institute
> Jupiter, 33458 FL, USA
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List <
[hidden email]> On
> Behalf Of Kilgore, Jason A.
> Sent: Montag, 6. August 2018 02:00
> To:
[hidden email]
> Subject: Re: Tetraspeck beads for PSFs **vendor reply**
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
>
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803533904&sdata=Y0ljWknnOSMcLyYNMumYrRqlgHuUT%2FxIZX3DVLVKSXQ%3D&reserved=0> Post images on
>
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803533904&sdata=r8t%2FK%2FWBRtBFvORGJumvdgBCSjQMfqCxdq2gPCUbYfM%3D&reserved=0> and include the link in your posting.
> *****
>
> That would make sense. You see, most of the Molecular Probes fluorescent
> microspheres utilize BODIPY dyes, which are quite lipophilic. Thus, it
> wouldn't surprise me that they might be drawn out by oils.
>
> Jason
>
>
> Jason A. Kilgore
> Technical Application Scientist
> Molecular Probes / EVOS Tech Support
> Thermo Fisher Scientific
>
> 1-800-955-6288 then option 4, then option 3, then option 2.
> Or dial direct at +1 541 335 0353
>
[hidden email]
>
> This communication is intended solely for the individual/entity to whom it
> is addressed. It may contain confidential or legally privileged
> information. Any unauthorized disclosure or copying is prohibited and may
> be unlawful. If you have received this communication in error, please
> notify the sender immediately and delete it from your system.
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of MODEL, MICHAEL
> Sent: Saturday, August 04, 2018 6:01 AM
> To:
[hidden email]
> Subject: Re: Tetraspeck beads for PSFs **vendor reply**
>
>
> Thanks to this discussion I realized what may be going on in our
> experiments as well. We are trying to image nonfluorescent Spherotech
> particles in slightly mismatched Cargille oils, and noticed that some oils
> produce something around the beads. I used to think it was trapped air but
> now understand that polystyrene gets partially dissolved in some of these
> oils.
>
>
> Mike
>
>
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of MODEL, MICHAEL <
[hidden email]>
> Sent: Saturday, August 4, 2018 8:36 AM
> To:
[hidden email]
> Subject: Re: Tetraspeck beads for PSFs **vendor reply**
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3D2sJQG16l24gxeA6DBeyaFYN7qoGvWnXfpGDTD5O-iDw%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803533904&sdata=bNkiqbGm7HiTn2k5URUafwGBDSkhED3wdp8Mhajr%2FYE%3D&reserved=0=>
> LISTSERV 16.0 - CONFOCALMICROSCOPY List at LISTS.UMN.EDU<
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3D2sJQG16l24gxeA6DBeyaFYN7qoGvWnXfpGDTD5O-iDw%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803533904&sdata=bNkiqbGm7HiTn2k5URUafwGBDSkhED3wdp8Mhajr%2FYE%3D&reserved=0=> >
> lists.umn.edu
>
[hidden email]: listserv archives. confocalmicroscopy
>
>
>
> Post images on
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3DdUrbX4VrVqWyfa4aWtxfj_D2EmsBl6Qx54Nr0kNb72M%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=Cm7zZcHxE0qEIeYJ5Rc4pBaOdE13giUMsUdZtlOausE%3D&reserved=0=> and include the link in your posting.
> *****
>
> I think you may simply keep dry beads on a coverslip and there is no need
> in a mounting medium. No spherical aberration would accumulate over a 0.1
> um depth
>
>
> Mike Model
>
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of Andreas Bruckbauer <
[hidden email]
> >
> Sent: Saturday, August 4, 2018 4:01 AM
> To:
[hidden email]
> Subject: Re: Tetraspeck beads for PSFs **vendor reply**
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3D2sJQG16l24gxeA6DBeyaFYN7qoGvWnXfpGDTD5O-iDw%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=TKeWG3t1Bywh0JoD6Ts24NyhQmBE8xJiqzqrWPwaVWw%3D&reserved=0=>
> LISTSERV 16.0 - CONFOCALMICROSCOPY List at LISTS.UMN.EDU<
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3D2sJQG16l24gxeA6DBeyaFYN7qoGvWnXfpGDTD5O-iDw%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=TKeWG3t1Bywh0JoD6Ts24NyhQmBE8xJiqzqrWPwaVWw%3D&reserved=0=> >
> lists.umn.edu
>
[hidden email]: listserv archives. confocalmicroscopy
>
>
>
> Post images on
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3DdUrbX4VrVqWyfa4aWtxfj_D2EmsBl6Qx54Nr0kNb72M%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=Cm7zZcHxE0qEIeYJ5Rc4pBaOdE13giUMsUdZtlOausE%3D&reserved=0=> and include the link in your posting.
> *****
>
> Thanks Jean for raising this issue, I have a similar problem, the dye
> diffuses out of the beads and leaves a large spot around them. It can
> happen very quickly when the slides get to warm and the current heatwave is
> not helping. Putting them in the fridge is also not an option because of
> drift once put back on the microscope.
> This is with mounting medium from Moleculr Probes for beads.
> No solution yet, I don't think it used to be like this, have you changed
> the beads?
>
> Aren't 100 nm beads too big for PSF measurements on the Elyra? 40 nm
> should be better.
>
> Best wishes
>
> Andreas
>
> Sent from my phone
>
> > On 3 Aug 2018, at 22:05, Kilgore, Jason A. <
>
[hidden email]> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furlde> > fense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-&
> > data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62
> > 063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&a
> > mp;sdata=2DlVDcjPD1Gczop%2B%2BnuaURDW%2B8r8vVmkPqIZKLMBpso%3D&rese
> > rved=0
> > 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=q6k2DsTcEGCcCb_WtVSz6hhI
> > l8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=7L
> > M14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8&s=2sJQG16l24gxeA6DBeyaFYN7qo
> > GvWnXfpGDTD5O-iDw&e= Post images on
> >
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DDwIF-g%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3D7LM14j0RRNKXNBbNA6rOihPlP1-soxMJ0ZS6FlUeWz8%26s%3DdUrbX4VrVqWyfa4aWtxfj_D2EmsBl6Qx54Nr0kNb72M%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=Cm7zZcHxE0qEIeYJ5Rc4pBaOdE13giUMsUdZtlOausE%3D&reserved=0=> and include the link in your posting.
> > *****
> >
> >
> > ** Vendor reply **
> >
> > Polystyrene microspheres, such as the TetraSpecks, will swell in the
> presence of certain solvents. So you'll want to use an aqueous-based
> mounting media, preferably one that cures (such as ProLong products or
> Fluoromount-G) instead of using an organic solvent-based one, like
> Cytoseal-60. When dye-labeled microspheres like these swell, the dye is
> free to diffuse out of the polystyrene matrix.
> >
> > Cheers,
> >
> > Jason
> >
> >
> > Jason A. Kilgore
> > Technical Application Scientist
> > Molecular Probes / EVOS Tech Support
> > Thermo Fisher Scientific
> >
> > 1-800-955-6288 then option 4, then option 3, then option 2.
> > Or dial direct at +1 541 335 0353
> >
[hidden email]
> >
> > This communication is intended solely for the individual/entity to whom
> it is addressed. It may contain confidential or legally privileged
> information. Any unauthorized disclosure or copying is prohibited and may
> be unlawful. If you have received this communication in error, please
> notify the sender immediately and delete it from your system.
> >
> >
> >
> > -----Original Message-----
> > From: Confocal Microscopy List
> > [mailto:
[hidden email]] On Behalf Of Jean Ross
> > Sent: Friday, August 03, 2018 12:49 PM
> > To:
[hidden email]
> > Subject: Tetraspeck beads for PSFs
> >
> > CAUTION: This email originated from outside of the organization. Do not
> click links or open attachments unless you recognize the sender and know
> the content is safe.
> >
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furlde> > fense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__lists.umn.edu_cgi-&
> > data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62
> > 063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&a
> > mp;sdata=2DlVDcjPD1Gczop%2B%2BnuaURDW%2B8r8vVmkPqIZKLMBpso%3D&rese
> > rved=0
> > 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=q6k2DsTcEGCcCb_WtVSz6hhI
> > l8hvYssy7sH8ZwfbbKU&r=MVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ&m=F8
> > IDbXdJNbnAT6ffW_HjXIuMlAG3s16jAZb0eldCiQU&s=_ngqUkuraSCI6zAmWzz6NEeoPt
> > sEIhu0PMk_Stm2Gko&e= Post images on
> >
>
https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.proofpoint.com%2Fv2%2Furl%3Fu%3Dhttp-3A__www.imgur.com%26d%3DDwIBaQ%26c%3Dq6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU%26r%3DMVp-2yJ1A-yQtCbZ-DE9xd0W6E7srQQpV-yioYjTLyQ%26m%3DF8IDbXdJNbnAT6ffW_HjXIuMlAG3s16jAZb0eldCiQU%26s%3DO2T49sxgjykKvUtqbrfKaSktZURKhrNNnPd3CQna8Gc%26e&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C4372055327a54fba751408d5fb62063d%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636691320803543916&sdata=YBRBOD%2Bwq%2BsKWcnPhHZXQYE8Bvm9p3xfUE1EmDjcOKo%3D&reserved=0=> and include the link in your posting.
> > *****
> >
> > Hi Everyone,
> >
> > I have been preparing 0.1um Tetraspeck bead slides to use when measuring
> point spread functions on our Zeiss Elyra PS1 super resolution microscope
> > and Zeiss LSM880 confocal with Airyscan. The problem is that over time
> > the beads appear to swell to about 3 times in size and lose much of
> > their fluorescence. I have tried many different mounting medias
> > (hardening and
> > nonhardening) all with the same result. Has anyone else experienced
> this problem? Any suggestions to solve the problem would be appreciated.
> >