Hello all. For those who are interested we carried out a relatively extensive study on the influence of bead size in the context of estimating the location of a bead by fitting a PSF. In short, we found that beads with a diameter of up to 100 nm yielded very similar results, suggesting that beads 100 nm or less in diameter would be suitable for PSF measurements. (https://doi.org/10.1371/journal.pone.0134112)

Best regards,
Jerry