Posted by
Ochoa, Lorenzo F. on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Safe-to-run-multiphoton-laser-at-100-power-for-overnight-continuous-imaging-tp7588726p7588728.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hello Dale,
I've ran a 2-P Mai Tai in the past for over 16 hours continuously at 1W power for a long imaging session. I did have a pockel cell to control the power that was actually getting to the sample but in terms of the laser being able to run for that long I had no issues. I haven't notice any damage to the optics so far either. The only problem I've ran into was a loss of communication between the laser and the computer with the control software. Hope that helps.
Lorenzo
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Moulding, Dale
Sent: Tuesday, September 25, 2018 7:56 AM
To:
[hidden email]
Subject: Safe to run multiphoton laser at 100% power for overnight continuous imaging?
WARNING: This email originated from outside of UTMB's email system. Do not click links or open attachments unless you recognize the sender and know the content is safe.
*****
To join, leave or search the confocal microscopy listserv, go to:
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Clfochoa%40UTMB.EDU%7Cae1aa8d9d79d4395517508d622e64662%7C7bef256d85db4526a72d31aea2546852%7C0%7C0%7C636734769755876244&sdata=aWnIWQvkLd0iWttfMxPFZvpV00g4Dtl%2FooBGCSHnJfc%3D&reserved=0Post images on
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Clfochoa%40UTMB.EDU%7Cae1aa8d9d79d4395517508d622e64662%7C7bef256d85db4526a72d31aea2546852%7C0%7C0%7C636734769755876244&sdata=DuOZDE%2Bq8uzh3qvVLiO7fRf4fGomhVAcV34ISyaTWwI%3D&reserved=0 and include the link in your posting.
*****
Dear list,
I've been asked to run our 2 photon laser (Mai Tai eHP DeepSee) at full power (3W) non-stop over night to scan a large sample.
When doing ablations we run at 100% and I hadn't even given it a second thought, as the ablations are over in a flash.
However, continuously (4 to 16 hours) running 100% 800nM at 3W I'm concerned we may damage the optics of the system. It will be on a Zeiss Examiner, going through Zeiss's 10x0.5 or 20x1.0 water dippers.
Has anyone ever imaged / ablated with a 2P laser continuously at full power?
Should I be concerned that the objectives may be damaged?
Cheers
Dale
Dale Moulding PhD FRMS
ICH Light Microscopy Facility
Room W2.06, BDRC office
UCL Great Ormond Street Institute of Child Health
30 Guilford St
London WC1N 1EH
Mob: 07787 699609
Tel: 020 7905 2753
https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ucl.ac.uk%2Fich%2Fcore-scientific-facilities-centres%2Fconfocal-microscopy&data=02%7C01%7Clfochoa%40UTMB.EDU%7Cae1aa8d9d79d4395517508d622e64662%7C7bef256d85db4526a72d31aea2546852%7C0%7C0%7C636734769755876244&sdata=HOf%2BtmkIwJftv7nQiwEyvII9EpbXKRRrX%2BwfpCwXgqo%3D&reserved=0