Re: Safe to run multiphoton laser at 100% power for overnight continuous imaging?

Posted by Moulding, Dale on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Safe-to-run-multiphoton-laser-at-100-power-for-overnight-continuous-imaging-tp7588726p7588732.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

Thanks for the replies. It will be at 800nM, so full power output, and will be continuously firing though the objective for at least 4 hours as it is not for imaging but for catalysis.
Zeiss aren't able to guarantee their objectives for this (drilling to the centre of the earth!), so if I do get a cracked lens / glue then I'm not sure our service contract will cover it.

@Lorenzo, for your 16 hour experiment were you sending the laser through the objective non-stop for the entire time, or just imaging periodically?
@Craig, agreed, it’s a lot of power, but not for imaging, for catalysis.
@Brian, I love the 10x0.5 dipper for 2P work. Massive improvement over the 10x0.3.
@Claire, I'm not sure Zeiss will cover the damage this time, and haven't really had a definitive answer, unless I've misunderstood them.

Cheers

Dale

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Craig Brideau
Sent: 25 September 2018 17:13
To: [hidden email]
Subject: Re: Safe to run multiphoton laser at 100% power for overnight continuous imaging?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

You do raise a valid point Brian, the power of most Ti:Saphs varies fairly dramatically by wavelength. The transmission through the optical system will also be different, and what the system considers '100%' may not actually be that much power under certain situations. In most cases, however, a Ti:Saph can produce much more power than is necessary for imaging.
Try measuring the power out of the objective when set to 100% to see what you are actually getting; maybe it isn't that much compared to the output of the laser once all the system losses are considered.

Craig

On Tue, Sep 25, 2018 at 9:41 AM Armstrong, Brian <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Dale, it depends in part on the wavelength. As I imagine you
> already know the output power at say 1050nm is far lower than say
> 800nm. I do not think the LASER "cares" what your output power is. As
> for the optics I cannot say. I think that they fall off in
> transmittance above a given wavelength in the IR range but I do not
> know if this is because of a coating or properties of the lens??? I
> have heard stories of people burning objectives. Pockels cells will also burn out after around 10 years or so.
>
> P.S. How do you like the 10x/0.5 W for 2P work?
>
> Brian Armstrong PhD
> Associate Research Professor
> Developmental and Stem Cell Biology
> Diabetes and Metabolic Diseases
> Director, Light Microscopy Core
> Beckman Research Institute, City of Hope
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]]
> On Behalf Of Moulding, Dale
> Sent: Tuesday, September 25, 2018 5:56 AM
> To: [hidden email]
> Subject: Safe to run multiphoton laser at 100% power for overnight
> continuous imaging?
>
> [Attention: This email came from an external source. Do not open
> attachments or click on links from unknown senders or unexpected
> emails.]
>
>
>
>
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear list,
>
> I've been asked to run our 2 photon laser (Mai Tai eHP DeepSee) at
> full power (3W) non-stop over night to scan a large sample.
> When doing ablations we run at 100% and I hadn't even given it a
> second thought, as the ablations are over in a flash.
> However, continuously (4 to 16 hours) running 100% 800nM at 3W I'm
> concerned we may damage the optics of the system. It will be on a
> Zeiss Examiner, going through Zeiss's 10x0.5 or 20x1.0 water dippers.
> Has anyone ever imaged / ablated with a 2P laser continuously at full
> power?
> Should I be concerned that the objectives may be damaged?
>
> Cheers
>
> Dale
>
>
> Dale Moulding PhD FRMS
> ICH Light Microscopy Facility
> Room W2.06, BDRC office
> UCL Great Ormond Street Institute of Child Health
> 30 Guilford St
> London WC1N 1EH
> Mob: 07787 699609
> Tel: 020 7905 2753
>
> http://www.ucl.ac.uk/ich/core-scientific-facilities-centres/confocal-m
> icroscopy
>
> ------------------------------------------------------------
> -SECURITY/CONFIDENTIALITY WARNING-
>
> This message and any attachments are intended solely for the
> individual or entity to which they are addressed. This communication
> may contain information that is privileged, confidential, or exempt
> from disclosure under applicable law (e.g., personal health
> information, research data, financial information). Because this
> e-mail has been sent without encryption, individuals other than the
> intended recipient may be able to view the information, forward it to
> others or tamper with the information without the knowledge or consent
> of the sender. If you are not the intended recipient, or the employee
> or person responsible for delivering the message to the intended
> recipient, any dissemination, distribution or copying of the
> communication is strictly prohibited. If you received the