Re: Safe to run multiphoton laser at 100% power for overnight continuous imaging?

Posted by Craig Brideau on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Safe-to-run-multiphoton-laser-at-100-power-for-overnight-continuous-imaging-tp7588726p7588739.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

In one case I noted the back aperture of an objective was partly melted
when the beam at high power overfilled the back aperture to an extreme
degree. If the back aperture of the lens is large and the beam diameter is
well matched then the majority of the power will go through the transparent
glass. It is clipping at the edges of apertures that leads to absorption
which can damage the lenses. It *is* possible for coatings on the lenses to
go even with a well aimed beam, but this is much less likely with
contemporary optical coatings.

Craig


On Wed, Sep 26, 2018 at 3:01 AM Christian Wilms <
[hidden email]> wrote:

> In over 15 years working with multiphoton systems (both in the lab and
> working for a microscopy company), I have not seen an objective 'burn out'.
> Pockels cells have upper limits, but the system should have been set up
> from the start to attenuate the beam prior to the Pockels, if the direct
> laser output is above that threshold. In general, Pockels cells can
> deteriorate over time, as Brian mentions, but that will heavily depend on
> how well the Pockels cell is aligned.
>
> Regarding your actual question: If you want a statement that you can rely
> on (including in terms of warrantee) I would strongly recommend asking your
> Zeiss representative.
>
> Cheers, Christian
>
> Dr. Christian Wilms / Research & Development Manager
> [hidden email] / +44 (0)1825 749933
> www.scientifica.uk.com
>
> > -----Original Message-----
> > From: Craig Brideau <[hidden email]>
> > Sent: 25 September 2018 17:13
> > Subject: Re: Safe to run multiphoton laser at 100% power for overnight
> > continuous imaging?
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > You do raise a valid point Brian, the power of most Ti:Saphs varies
> fairly
> > dramatically by wavelength. The transmission through the optical system
> will
> > also be different, and what the system considers '100%' may not actually
> be
> > that much power under certain situations. In most cases, however, a
> Ti:Saph
> > can produce much more power than is necessary for imaging.
> > Try measuring the power out of the objective when set to 100% to see what
> > you are actually getting; maybe it isn't that much compared to the
> output of
> > the laser once all the system losses are considered.
> >
> > Craig
> >
> > On Tue, Sep 25, 2018 at 9:41 AM Armstrong, Brian <[hidden email]>
> > wrote:
> >
> > > *****
> > > To join, leave or search the confocal microscopy listserv, go to:
> > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > > Post images on http://www.imgur.com and include the link in your
> posting.
> > > *****
> > >
> > > Hi Dale, it depends in part on the wavelength. As I imagine you
> > > already know the output power at say 1050nm is far lower than say
> > > 800nm. I do not think the LASER "cares" what your output power is. As
> > > for the optics I cannot say. I think that they fall off in
> > > transmittance above a given wavelength in the IR range but I do not
> > > know if this is because of a coating or properties of the lens??? I
> > > have heard stories of people burning objectives. Pockels cells will
> also burn
> > out after around 10 years or so.
> > >
> > > P.S. How do you like the 10x/0.5 W for 2P work?
> > >
> > > Brian Armstrong PhD
> > > Associate Research Professor
> > > Developmental and Stem Cell Biology
> > > Diabetes and Metabolic Diseases
> > > Director, Light Microscopy Core
> > > Beckman Research Institute, City of Hope
> > >
> > >
> > >
> > > -----Original Message-----
> > > From: Confocal Microscopy List
> > > [mailto:[hidden email]]
> > > On Behalf Of Moulding, Dale
> > > Sent: Tuesday, September 25, 2018 5:56 AM
> > > To: [hidden email]
> > > Subject: Safe to run multiphoton laser at 100% power for overnight
> > > continuous imaging?
> > >
> > > [Attention: This email came from an external source. Do not open
> > > attachments or click on links from unknown senders or unexpected
> > > emails.]
> > >
> > >
> > >
> > >
> > >
> > > *****
> > > To join, leave or search the confocal microscopy listserv, go to:
> > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > > Post images on http://www.imgur.com and include the link in your
> posting.
> > > *****
> > >
> > > Dear list,
> > >
> > > I've been asked to run our 2 photon laser (Mai Tai eHP DeepSee) at
> > > full power (3W) non-stop over night to scan a large sample.
> > > When doing ablations we run at 100% and I hadn't even given it a
> > > second thought, as the ablations are over in a flash.
> > > However, continuously (4 to 16 hours) running 100% 800nM at 3W I'm
> > > concerned we may damage the optics of the system. It will be on a
> > > Zeiss Examiner, going through Zeiss's 10x0.5 or 20x1.0 water dippers.
> > > Has anyone ever imaged / ablated with a 2P laser continuously at full
> > > power?
> > > Should I be concerned that the objectives may be damaged?
> > >
> > > Cheers
> > >
> > > Dale
> > >
> > >
> > > Dale Moulding PhD FRMS
> > > ICH Light Microscopy Facility
> > > Room W2.06, BDRC office
> > > UCL Great Ormond Street Institute of Child Health
> > > 30 Guilford St
> > > London WC1N 1EH
> > > Mob: 07787 699609
> > > Tel: 020 7905 2753
> > >
> > > http://www.ucl.ac.uk/ich/core-scientific-facilities-centres/confocal-m
> > > icroscopy
> > >
> > > ------------------------------------------------------------
> > > -SECURITY/CONFIDENTIALITY WARNING-
> > >
> > > This message and any attachments are intended solely for the
> > > individual or entity to which they are addressed. This communication
> > > may contain information that is privileged, confidential, or exempt
> > > from disclosure under applicable law (e.g., personal health
> > > information, research data, financial information). Because this
> > > e-mail has been sent without encryption, individuals other than the
> > > intended recipient may be able to view the information, forward it to
> > > others or tamper with the information without the knowledge or consent
> > > of the sender. If you are not the intended recipient, or the employee
> > > or person responsible for delivering the message to the intended
> > > recipient, any dissemination, distribution or copying of the
> > > communication is strictly prohibited. If you received the
> > > communication in error, please notify the sender immediately by
> > > replying to this message and deleting the message and any accompanying
> > > files from your system. If, due to the security risks, you do not wish
> > > to receive further communications via e-mail, please reply to this
> message
> > and inform the sender that you do not wish to receive further e-mail from
> > the sender.
> > > (LCP301)
> > > ------------------------------------------------------------
> > >
>