Re: thermal imaging through microscope

Posted by Craig Brideau on
URL: http://confocal-microscopy-list.275.s1.nabble.com/thermal-imaging-through-microscope-tp7588813p7588815.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

BK7 optical glass, which is pretty common in microscope optics, drops off
fairly sharply in transmission by 2 to 2.5um, so that will be a problem.
The last time I heard about a lab attempting thermal imaging they had a
custom gold-coated reflective objective constructed for the purpose, and
designed all their other optics with IR glass like zinc selenide or
similar. It seems unlikely this would work at all through a conventional
microscope. I can point you to some additional resources if you are
interested.

Craig

On Fri, Oct 12, 2018 at 12:17 PM Kathryn Spencer <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi;
>                 I'm looking for suggestions and equipment to do thermal
> imaging through a microscope. We want to image C. elegans to look at their
> body temperature under different metabolic conditions. While I know they
> are poikilotherms, we believe they will show a difference to their local
> background under these conditions. Can you recommend a decent camera for
> thermal imaging? Are there IR cutoff filters or optics that will not
> transmit IR wavelengths in a basic fluorescence-type microscope (needed for
> worm-sized resolution ~10x mag)?
>                 Thanks in advance.