Re: Problems with ZEISS ZEN Black after Windows 7 updates this week

Posted by Ariel, Pablo on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Problems-with-ZEISS-ZEN-Black-after-Windows-7-updates-this-week-tp7589022p7589035.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Roland,

Thanks for providing this valuable information to the community; you just saved my staff and I a lot of time.

I can confirm that KB 4480970 (which is labeled a "security update") caused startup problems on both an LSM700 and LSM710 system, both running Zen Black on Windows 7. The errors were different in each case, but made the software unusable for both microscopes. Uninstalling the update fixed the issue. I have not tried reinstalling.

Thanks again,
Pablo


Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory
Brinkhous-Bullitt Bldg B04
Department of Pathology and Laboratory Medicine
University of North Carolina at Chapel Hill
http://www.med.unc.edu/microscopy
Tel: 919-966-2413




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Roland Nitschke
Sent: Friday, January 11, 2019 6:27 AM
To: [hidden email]
Subject: Problems with ZEISS ZEN Black after Windows 7 updates this week

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

many of you might have problems in running ZEISS ZEN Black systems (online or offline) after a bunch of the last Windows 7 updates this week.
Problems are related to the Windows updates with the numbers

# KB 4480970
# KB 4480960

We had success in getting a running ZEN BLACK system again after de-installing them and rebooting.

On most systems we also see that we can even reinstall them afterwards again and reboot the system and ZEN still is running.

If it does not work you have to block the Windows update in the meanwhile, until there is a fix from Microsoft.

This is not logic, but its Windows.....


Best wishes and good luck

Roland Nitschke

Dr. Roland Nitschke Albert-Ludwigs University Freiburg Life Imaging Center (LIC) in ZBSA Microscopy and Image Analysis Platform (MIAP)