Hi Silvie,

That's useful approach to try. Whenever possible I try an individual
approach matching the scope and depth of the material to the trainee's
background and skill level. The worst scenario is when a lab sends in
several trainees of different level (e.g. student, technician and
postdoc). Do you accept several trainees for the same session?
Next major difficulty  for the trainee that I noticed is the transition
from imaging training sample to the self-prepared one. Here the ability
to duplicate tutorial actions usually is not enough , in depth
understanding is required, and of course preparing quality samples.

Best regards,
Arvydas
>>> Sylvie Le Guyader <[hidden email]> 10/16/2019 9:54 AM >>>
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=1I4RBPK585IqkSCgWSgQlrCupPknMSbFrCOaHZSqDEk&e=


Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=Xzk8Ro7YdVBAzZ37NqwqGMI3eVf5cVL34lfiz_pL-ZY&e=
 and include the link in your posting.

*****

’ significant fraction of the basic confocal trainees bored after 1-2
hrs of total 3.’
Definitely! That would bore me too and I am not from the millennium
generation! ��
I believe that the only way is to make very short videos and give
quizzes half way or just afterwards. E.g. our trainee watches our video
about bit depth, saturation and under exposure knowing that right after
the video we will ask to show us where the buttons are and if there was
any saturation/under exposure in the image they took immediately before
watching the video. They are encouraged to stop the video and push the
buttons on the software whenever they feel like it. ��

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Blickagången 16,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI
website<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_bionut_welcome-2Dto-2Dthe-2Dlci-2Dfacility&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=eHHGZSUe6-p2KrVzjC_VHcLg3gh9nGLmDr6tud4_KTE&e=
>

Follow our microscopy
blog!<https://urldefense.proofpoint.com/v2/url?u=http-3A__microscopykarolinska.se_&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=4wwhhmKRfu8gO9SaXZJKMAf72Q8Ng7Jhnh3SLgiIFn4&e=
>


From: Arvydas Matiukas <[hidden email]>
Sent: 16 October 2019 13:53
To: Sylvie Le Guyader <[hidden email]>;
[hidden email]
Subject: Re: [EXTERNAL] Re: Confocal and Widefield microscopy training
material

Hello Sylvie,

Thanks for pointing to these video tutorials by Alyona Minina. They
nicely complement traditional powerpoint material and hopefully could
save some my time. I am curious to check how much longer video material
keeps trainee's attention focussed, especially for millenial youtube
generation. I noticed that traditional verbal and powerpoint
presentation makes significant fraction of the basic confocal trainees
bored after 1-2 hrs of total 3.

with best wishes Arvydas




+++++++++++++++++++++++++++++
Arvydas Matiukas, Ph.D.
Manager of NRB Shared Equipment
Director of Confocal&Two-Photon Core
SUNY Upstate Medical University

>>> Sylvie Le Guyader
<[hidden email]<mailto:[hidden email]>> 10/16/19 6:28
AM >>>
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=eFtCsX4RwhV6lbXscD96MNTyssit68QdB4fZA95hrQQ&e=

Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=fAjOkG_slD2KBsdz68xI4iL4ibgq5pezOL5F3lRr2y4&e=
and include the link in your posting.

*****

Hi Niyanta



Alyona Minina has tons of videos on you tube. Check her
channel<https://urldefense.proofpoint.com/v2/url?u=https-3A__www.youtube.com_channel_UCfRx0TPF-2DR1TKYrPPu7OwWA_videos&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=o-JPkTgkyERoEbgIwRA5ZY6W1TvsWK5gsmDv1ge4toE&e=
<https://urldefense.proofpoint.com/v2/url?u=https-3A__www.youtube.com_channel_UCfRx0TPF-2DR1TKYrPPu7OwWA_videos&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=o-JPkTgkyERoEbgIwRA5ZY6W1TvsWK5gsmDv1ge4toE&e=%20>
>.




Med vänlig hälsning / Best regards



Sylvie



@@@@@@@@@@@@@@@@@@@@@@@@

Sylvie Le Guyader, PhD

Live Cell Imaging Facility Manager

Karolinska Institutet- Bionut Dpt

Blickagången 16,

Room 7362 (lab)/7840 (office)

14157 Huddinge, Sweden

mobile: +46 (0) 73 733 5008

LCI website

Follow our microscopy blog!



-----Original Message-----
From: Confocal Microscopy List
<[hidden email]<mailto:[hidden email]>>
On Behalf Of Niyanta Kumar
Sent: 11 October 2019 22:00
To:
[hidden email]<mailto:[hidden email]>
Subject: Confocal and Widefield microscopy training material



*****

To join, leave or search the confocal microscopy listserv, go to:

https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Flists.umn.edu-252Fcgi-2Dbin-252Fwa-253FA0-253Dconfocalmicroscopy-26amp-3Bdata-3D02-257C01-257Csylvie.le.guyader-2540KI.SE-257C87a9cd3e2b2d42366e3808d74e8707d5-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637064214752285084-26amp-3Bsdata-3DXtByjWY8X-252BLYVFhPt7T7XqnizH6v9I0RZalnJIYo23s-253D-26amp-3Breserved-3D0&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=9ndwMylZEtR6J8pLSr5vVXCqk8bBKkmliAmW2jootR0&e=


Post images on
https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.imgur.com-26amp-3Bdata-3D02-257C01-257Csylvie.le.guyader-2540KI.SE-257C87a9cd3e2b2d42366e3808d74e8707d5-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637064214752285084-26amp-3Bsdata-3D6m3Wpvhetv-252BypOIa61Jr8jZeKn0olhUe50y8LbuvJho-253D-26amp-3Breserved-3D0&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=pZwg3cOxtPEB45-6ZJuKjpBvaR4hMvAION4xiM1-TXU&e=
and include the link in your posting.


*****



Hi all,

Can someone suggest any existing resources and formats to pull from for
designing new and advanced user trainings for confocal and widefield
microscopy? I need to design trainings for a Zeiss LSM 980 Airyscan 2
confocal and an Axio Obsever widefield scope. Interactive resources or
existing quizzes that can ensure users know what they need to would also
be helpful.

Thanks,

Niyanta



När du skickar e-post till Karolinska Institutet (KI) innebär detta att
KI kommer att behandla dina personuppgifter. Hr finns information om hur
KI behandlar
personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=EQs1kTVGSWHn1Bbnwf8tsm9ZmeiX9rnpVF3Gx8njx2g&e=
<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=EQs1kTVGSWHn1Bbnwf8tsm9ZmeiX9rnpVF3Gx8njx2g&e=%20>
>.



Sending email to Karolinska Institutet (KI) will result in KI
processing your personal data. You can read more about KI’s processing
of personal data
here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=3Bvbvriv0Qsm1rvTGdCTn69BFi2whwW0uFp_0iJEm9g&e=
<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=fqPKyH7OvDC2YG-CgHu2VMvZCDwOwF5ALyEnXO-YnB8&s=3Bvbvriv0Qsm1rvTGdCTn69BFi2whwW0uFp_0iJEm9g&e=%20>
>.




När du skickar e-post till Karolinska Institutet (KI) innebär detta att
KI kommer att behandla dina personuppgifter. Här finns information om
hur KI behandlar
personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=8j7d7fhNJQoj3Rai2nIAydL3Me-74TRwHI-43to-b0M&e=
>.



Sending email to Karolinska Institutet (KI) will result in KI
processing your personal data. You can read more about KI’s processing
of personal data
here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIGaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=XGh1Uu87TWxekpsyPOgE9NKlXFtrItrNJpVgKlIIhEk&s=qzKolo6QFEOyQO9RlKOPvFp7_-GjfPYp-nmKl8nobas&e=
>.