Hi Claire,
Just curious - were the measurements done in water?
Regards,
Scott
-----------------------
Scott Henderson, Ph.D.
Professor, Dept. Molecular Medicine
Director, TSRI Microscopy Core
Room MB-22
Scripps Research
10550 North Torrey Pines Rd.
La Jolla, CA 92037
858-784-8163
On Apr 17, 2020, at 1:23 PM, Claire Brown, Dr. <
[hidden email]<mailto:
[hidden email]>> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hi All,
We did some extensive intensity calibration experiments with different intensity green and red microspheres.
The green microspheres were 3.7% and 35% relative intensities with an intensity ratio of 10.6. We measured the intensity ratio with lots of different microscopes and lots of different lenses and very consistently go a ratio of 8.
I can't seem to figure this out at all. It means the bright beads have to be a bit dimmer than expected or the dim beads a bit brighter than expected.
The calibration would have been done by ThermoFisher - I would guess they do this by flow cytometry. Maybe it could just be that microscopy measures the intensities more accurately? Flow just gets on data point per sphere? Maybe the bright beads bleach relatively more than the dim ones when you are imaging on a CLSM?
Any ideas are welcome!
All the best,
Claire