Hi Xavi,

For fastest photobleaching, generally best to use maximum power for as
brief a time as possible. The more photons, the bigger the spot. Oxygen
radicals and dye radicals diffuse short distances relative to PSF
(nanometers ... though activated tyramide radicals have a diffusion
radius of ~100 nm, and can be restrained by increasing viscosity or
additives ... cytoplasm is higher viscosity and density than water ...
so: radicals diffusion radius might be suppressed in live cells that are
[over]expressing catalase).

A(nother) good PSF calculator is available free at SVI

https://svi.nl/NyquistCalculator

Microscopy Nyquist rate and PSF calculator.

I liked this recent publication on the photophysics and photochemistry
of fluorescence - and generalization across the UV-Visible fluorophores:

Aleksandr Barulin and Jeroome Wenger 2020
Ultraviolet Photostability Improvement for Autofluorescence Correlation
Spectroscopy on Label-Free Proteins
J. Phys. Chem. Lett. 2020, 11, 2027−2035
https://dx.doi.org/10.1021/acs.jpclett.0c00209
 From abstract: "Our results show that the underlying photochemical
concepts initially derived for organic visible fluorescent dyes are
quite general."

George

p.s. tyramide references:

* Chen ... Belmont 2018 J Cell Biol https://doi.org/10.1083/jcb.201807108

* Phil Moens, pers. comm. [1993?] (ex-NEN, before acquisition by
PerkinElmer and Phenoptics division spinout to Akoya).

On 4/23/2020 6:21 PM, SANJUAN SAMARRA, XAVIER wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear listers,
>
> When using the Bleach Point feature in the Leica confocal systems (I'm sure
> it also exists in the other brands), I don't know how to measure the
> surface of the bleached area. I figure out the wavelength of the laser beam
> and the NA of the lens affect the size of that area, so would it be correct
> to use the resolution formula to calculate it?  Or am I mixing up things
> here?
>
> Thank you very much,
>
> Xavi.
>
> ___________________________________
>
> *Xavier Sanjuan*
> Advanced Light Microscopy Unit
>
> Parc de Recerca Biomèdica de Barcelona
> Doctor Aiguader, 88
> 08003 Barcelona - Spain
> Tel: + 34 93 316 0195 (ext 1195 dins el PRBB)
> Fax: + 34 93 316 09 01
> E-mail: [hidden email]
> Web:
> http://www.crg.eu/en/core/programmes-groups/advanced-light-microscopy-unit