In case it helps, I was just having RTC connection issues on an LSM 710 and
it turns out the RTC was not booting up correctly for lack of a keyboard
(we think BIOS was reset). Connecting a PS/2 keyboard to the back of the
RTC fixed the issue while Zeiss resets the BIOS and replaces the
motherboard battery.

-Esteban

On Sun, Jun 28, 2020 at 12:15 PM Jacqueline Ross <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Good morning,
>
> I don't know if this will help but writing to you just in case the problem
> is less serious than you think.
>
> We have a Zeiss LSM 710. After we moved to Windows 7 some years ago, we
> had to change the order in which we turned things on. If we turn on the RTC
> before the PC, we get the same error. Using the Reset button doesn't work,
> we just have to turn off the RTC, exit ZEN, restart the RTC and wait for
> the start up sounds  Then restart ZEN and all is OK.
>
> We changed our practice to turning on the RTC after the PC is on and after
> we have logged on. This solved the problem.
>
> Also our system has a back up copy of the system on a USB that is stored
> inside the PC. I think this is standard for Zeiss so you should also have
> this.
>
> Kind regards,
>
> Jacqui
> ________________________________________
> From: Confocal Microscopy List [[hidden email]] on
> behalf of Prof. Dr. K. S. Krishnamurthy [[hidden email]]
> Sent: 28 June 2020 16:52
> To: [hidden email]
> Subject: Failed RTC connection
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly
> for polarization studies on liquid crystals. Recently, the Zen software
> failed to launch with "fail to connect to RTC..." message. Resetting did
> not work. On the monitor, however, 'Real time connection connected..'
> message appears. It is suspected that the IDE, transcend 128 MB flash
> module, used to boot may have been corrupted. If a new IDE flash module  is
> purchased, the program data will have to be written into it. Is there a way
> of getting the data? Since the model is out of service, local Zeiss rep has
> expressed inability to extend technical support. It seems such a pity that
> for the sake of the IDE, the instrument should remain defunct.  We would be
> indebted for any suggestions to bring the confocal onroad.
>