For a DMD I would definitely recommend using an LED over a laser. Lasers
result in some pretty nasty speckle on the image plane, and also cause a
sort of blazing effect on the DMD, resulting in the illumination on the
image plane being quite non-uniform. You can also get a 1W 405 nm LED for
pretty cheap:
https://www.digikey.com/product-detail/en/new-energy/LST1-01G01-UV04-00/1672-1126-ND/7926553If you really want some top range oomph, you can go up to 16W still for
much less than a laser:
https://www.digikey.com/product-detail/en/luminus-devices-inc/CBM-120-UV-C31-K385-21/1214-1324-ND/5719784Here is a picture of the 16W LED in action; you definitely want eye
protection around this thing:
https://drive.google.com/file/d/1A639akV9E1MVhoqsdxKO3M-oktaHPkHX/Cheers,
Ben Smith
On Mon, Sep 21, 2020 at 7:11 AM Antonio Jose Pereira <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all,
>
> I'm looking for a 0.5-1W , 405nm, can have many spatial as well as
> longitudinal modes, 405nm laser. This is to feed a photoconversion module
> (Mosaic), a DMD-based system that wastes most power. That's why we need
> such high power source.
>
> I appreciate if you could give suggestions on low-budget options, which
> I'm failing to find ...
>
> Thank you so much,
> Antonio
>
>
>
> Antonio Pereira
> CID lab
> i3S/IBMC, Universidade do Porto
> Room 001.S2B, +351 22 607 49 59 Ext. 6127
>
--
Benjamin E. Smith, Ph. D.
Imaging Specialist, Vision Science
University of California, Berkeley
195 Life Sciences Addition
Berkeley, CA 94720-3200
Tel (510) 642-9712
Fax (510) 643-6791
e-mail:
[hidden email]
https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/