Dear All,
Thank you very much for so many responses, although the discussion went a bit into a direction that I did not anticipate ??
But yes, it is all about preventing oil from getting into the microscope. Our 40x/1.3 Fluar lens takes a lot of oil and I frequently have to clean its bottom side from oil. In contrast, I never saw any oil on the back of 63x/1.4 even though it is used as frequently as the 40x.
I have to admit I have been skeptical about hair elastics, because when they saturate with oil they will just let it go. IMHO, it is replacing "oil cleaning" approach with "oil buffering" approach. The "cleaning" approach has worked for us for many years. It failed when an unexperienced user (who apparently did not take seriously what I was talking during the training) made two mistakes: did not clean the lens and let it stay in the optic path after the work instead of changing for 10x. In addition, it was a holiday time, the system was not used for the next few days, giving the oil plenty of time. So, it was a coincidence of several factors, but it can happen again.
I think, hearing so many positive opinions about hair elastics, I will give them the 2nd chance in our facility. Aqua-Stop in my opinion is overpriced, and we do not currently do any experiments under perfusion to justify such purchase (this may change in the future). However, I would not like to give up on the protective window strategy. I think placing a glass element in the infinity space is not unusual. It would function as the last stronghold against the oil. And it would be considerably cheaper than Aqua-Stop. I would of course test PSF. If PSF is not affected, I guess the protective window is OK with imaging. So, if you still have some advice which one would be most suitable, I would appreciate it.
Best,
Tomasz
Tomasz Wegierski, PhD
International Institute of Molecular and Cell Biology
Trojdena 4, 02-109 Warsaw, POLAND
tel: +48-22 597 0763
fax: +48 22 597 0715
http://www.iimcb.gov.pl/________________________________
From: Confocal Microscopy List <
[hidden email]> on behalf of Frohlich, Victoria <
[hidden email]>
Sent: Tuesday, October 27, 2020 4:38 PM
To:
[hidden email] <
[hidden email]>
Subject: Re: Protective window for a confocal microscope?
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Just my 2 cents.
Scrunchies work well but are not fool-proof. Cotton not polyester material are more absorbent.
In addition to scrunchies I suggest wrapping the lens barrel near the nose of the objective with plumbers tape to prevent oil from seeping into the barrel. It can be done so that the lens retraction still functions. Then once the lens is on the nosepiece of the scope the plumbers tape can be wrapped there as well.
I've also used thin rubber sheets or parafilm sheets, as well as filter paper (thanks Mike S.!), to draw oil away from the lens and nosepiece.
Best method, however, is to teach users proper oil application and ensure they clean up after themselves.
Regards,
Vickie
Victoria Centonze Frohlich, PhD
Director, Light Microscopy Shared Resource
Cell & Tissue Imaging Center
St. Jude Children’s Research Hospital
262 Danny Thomas Place
MS 312
Memphis, TN 38105
Office: D1052D
Phone and Digital Fax: 1-901-595-2536
Cell: 1-901-907-2596
Paper Fax: 1-901-595-2909
Email:
[hidden email]
CTIC-LM Location: D1055 and D1056
Facility Phone: 1-901-595-3439
http://home.stjude.org/cell-tissue-imaging/Pages/default.aspxFacility Acknowledgement for publications:
“Images were acquired at the Cell & Tissue Imaging Center which is supported by SJCRH and NCI P30 CA021765.”
-----Original Message-----
From: Confocal Microscopy List <
[hidden email]> On Behalf Of Michael Stanley
Sent: Tuesday, October 27, 2020 8:19 AM
To:
[hidden email]
Subject: Re: Protective window for a confocal microscope?
Caution: External Sender
*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573449006%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=WiaypCuZUL5s03B%2Br1rcmMLyOazUGYLj%2BmW7hRFcyIU%3D&reserved=0Post images on
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573449006%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=DKQAcpfaPdTDony5B7X%2Fdi%2BCVChjA7voJxvlDKq9suo%3D&reserved=0 and include the link in your posting.
*****
in the 'old days' in the core we used to cut filter paper and then bore a hole with a cork-bore slightly smaller than the lens. slide 2-3 (or as many as will fit) layers of the filter paper over the objective. this will not stop a "flood" of oil (which sounds very possible for some) but it will hold much more than you would think. and, the most important feature of this little trick is that the oil will show very quickly on the paper and hopefully will trigger a response.!
some of the darker filter papers will show the oil better, but this will vary by manufacturer.
no commercial interest here, just trying to protect all the optics in the system.
michael
c. michael stanley, phd
senior applications scientist
chroma technology corp.
10 imtec lane
bellows falls, vt 05101
[hidden email]
[hidden email]
-----Original Message-----
From: Confocal Microscopy List <
[hidden email]> On Behalf Of Gary Laevsky
Sent: Tuesday, October 27, 2020 10:19 AM
To:
[hidden email]
Subject: Re: Protective window for a confocal microscope?
*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573449006%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=WiaypCuZUL5s03B%2Br1rcmMLyOazUGYLj%2BmW7hRFcyIU%3D&reserved=0Post images on
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573459000%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=yhWCT39l%2FiZJ9i8H0f4EM7MIstxliSvBVsjVQVNN8cI%3D&reserved=0 and include the link in your posting.
*****
The hair ties work great with objectives that have enough of a neck for the tie to sit on, next to the barrel.
Our 40X objective (most frequently used) has no/very small neck, so the hair ties don't work.
I fairly regularly (bi-weekly) have to take the objective off, store it in it's case upside down, and put it in a 37C oven. Ridiculous, in a bad way, how much oil comes out. The 40X also doesn't have "lip" that acts as a bit of a reservoir (unlike the 60 and 100).
On one of my heavily used systems, I have to go so far as to dismantle the top (stage, objective turret, and fluorescent turret) to get to the piece of glass that prevents leakage into the body (amazingly awesome feature of the stand). Takes me 15 minutes. Although I shouldn't have to do that ...
On Tue, Oct 27, 2020 at 9:13 AM Cammer, Michael <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists> .umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=04%7C01%7Cv
> ictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C223
> 40fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573459000%7CUnknown%7
> CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXV
> CI6Mn0%3D%7C1000&sdata=q2oFBgmsbEfra%2BIBWB6i8xwEJ%2B8SF%2F59WSo3Y
> %2BKvFKg%3D&reserved=0 Post images on
>
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573459000%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=yhWCT39l%2FiZJ9i8H0f4EM7MIstxliSvBVsjVQVNN8cI%3D&reserved=0 and include the link in your posting.
> *****
>
> Aquastop works well for big spills when installed correctly and has
> proven effective on some of our Zeiss scopes. But we did have a user
> who spilled over or around it on our Zeiss 710, so it is not foolproof.
>
>
> Hair elastics are great for the oil.
>
>
> Michael Cammer, Sr Research Scientist, DART Microscopy Laboratory
>
> NYU Langone Health, 540 First Avenue, SK2 Microscopy Suite, New York,
> NY
> 10016
>
>
[hidden email]<mailto:
[hidden email]>
>
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fnyulm> c.org%2Fmicros&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b8
> 9a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7
> C0%7C637394093573459000%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLC
> JQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=Ud4O3pAe
> 6c4HlzFgcjkQBvmAAz45ILNKOj8mnQScAsg%3D&reserved=0
>
https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fmicro> scopynotes.com%2F&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C
> 6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C
> 0%7C0%7C637394093573459000%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDA
> iLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=e6srs
> AN7BHwtFBKtNbXavROHTyX2m0CdnPe3l7VnUPA%3D&reserved=0
>
> Voice direct only, no text or messages: 1-914-309-3270 and
> 1-646-501-0567
>
>
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of Doug Richardson <
[hidden email]>
> Sent: Tuesday, October 27, 2020 7:41:45 AM
> To:
[hidden email]
> Subject: Re: Protective window for a confocal microscope?
>
> [EXTERNAL]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-> 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFBA&c=j5oPpO0eBH1iio48DtsedeEl
> Zfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=ke
> g1PHlpRuZ6hHabv7CvVJ2rsWSRa_2pkARNBtS-dFQ&s=k-eCFZaqV2J3Nbh8c16znCG7uP
> Pgwni6bBeSRyqYRAE&e=
> Post images on
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=Dw> IFBA&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC5
> 0tp7rW2nBkvV7fujQf0RknE5bU&m=keg1PHlpRuZ6hHabv7CvVJ2rsWSRa_2pkARNBtS-d
> FQ&s=o9KAzzxXBHkEPM_YiZYo9mb7m2x6GjWRoUHh-BzcNSg&e=
> and include the link in your posting.
> *****
>
> I would highly recommend the aquastop system from Zeiss. It is
> overpriced for some pieces of plastic, but it forms a very tight seal
> and pays for itself many times over. It is far superior to hair
> elastics/glove fingers/condoms which we used in the past.
>
> Doug
>
> Get Outlook for Android
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of Tomek Węgierski <
[hidden email]>
> Sent: Tuesday, October 27, 2020 7:12:10 AM
> To:
[hidden email]
> <
[hidden email]>
> Subject: Protective window for a confocal microscope?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-> 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFBA&c=j5oPpO0eBH1iio48DtsedeEl
> Zfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=ke
> g1PHlpRuZ6hHabv7CvVJ2rsWSRa_2pkARNBtS-dFQ&s=k-eCFZaqV2J3Nbh8c16znCG7uP
> Pgwni6bBeSRyqYRAE&e=
> Post images on
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=Dw> IFBA&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC5
> 0tp7rW2nBkvV7fujQf0RknE5bU&m=keg1PHlpRuZ6hHabv7CvVJ2rsWSRa_2pkARNBtS-d
> FQ&s=o9KAzzxXBHkEPM_YiZYo9mb7m2x6GjWRoUHh-BzcNSg&e=
> and include the link in your posting.
> *****
>
> Dear All,
> I am considering a purchase of a protective window for our Zeiss
> LSM800 and I would like to hear your advice whether this is a good
> idea, and if yes, which one to buy.
> The reason is our LSM800 is on an inverted stand and we already got a
> leakage of oil once which went down onto Optovar. And this was despite
> my regular checks of objectives (both sides) for leaking oil and
> extensive education of users on how important it is not to use too
> much oil and to clean lenses, especially on inverted stands.
> Zeiss Observer has a dummy slider with an opening for 32 mm filters. I
> hope, a protective window there should stop the oil, if the leakage
> happens again.
> 1) do you think placing a protective window in a confocal microscope
> is a good idea or not really, because the imaging will be affected?
> This would be in infinity space, so I guess it should not be
> detrimental....
> 2) which characteristics of protective window I should take into
> consideration (substrate, thickness, coating, flatness, surface
> quality, parallelism)? Currently I am inclined into protective glasses
> by Edmund Optics such as 1-mm thick MgF2-coated Sapphire window, which
> has high transmission, or 2-mm thick MgF2-coated B270 window, which
> has a bit worse transmission but better surface parameters (and is
> much cheaper).
> best regards,
> Tomasz
>
>
> Tomasz Wegierski, PhD
> International Institute of Molecular and Cell Biology Trojdena 4,
> 02-109 Warsaw, POLAND
> tel: +48-22 597 0763
> fax: +48 22 597 0715
>
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.iimcb.gov.pl_&> d=DwIFBA&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDi
> PlC50tp7rW2nBkvV7fujQf0RknE5bU&m=keg1PHlpRuZ6hHabv7CvVJ2rsWSRa_2pkARNB
> tS-dFQ&s=u7qHptX43byvVfMe-Zn04ev5KF4-ToW1ss01pq0eVtM&e=
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of
> the intended recipient(s) and may contain information that is
> proprietary, confidential, and exempt from disclosure under applicable
> law. Any unauthorized review, use, disclosure, or distribution is
> prohibited. If you have received this email in error please notify the
> sender by return email and delete the original message. Please note,
> the recipient should check this email and any attachments for the
> presence of viruses. The organization accepts no liability for any
> damage caused by any virus transmitted by this email.
> =================================
>
--
Best,
Gary Laevsky, Ph.D.
Director, Confocal Imaging Facility
Nikon Center of Excellence
Co-Founder, North Atlantic Microscopy Society (NAMS)
https://nam11.safelinks.protection.outlook.com/?url=https%3A%2F%2Fnamsmicroscopy.com%2F&data=04%7C01%7Cvictoria.frohlich%40STJUDE.ORG%7C6b89a1a0f4cc42f5937e08d87a8d1047%7C22340fa892264871b677d3b3e377af72%7C0%7C0%7C637394093573459000%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=7foXnZ0OuaU32Fj2dlCniXz%2Bn8iWTBvMyOkGj0rTX10%3D&reserved=0 Dept. of Molecular Biology Washington Rd.
Princeton University
Princeton, New Jersey, 08544-1014
(O) 609 258 5432
(C) 508 507 1310
________________________________
Email Disclaimer: www.stjude.org/emaildisclaimer<
http://www.stjude.org/emaildisclaimer>
Consultation Disclaimer: www.stjude.org/consultationdisclaimer<
http://www.stjude.org/consultationdisclaimer>