Re: Zeiss oil on Leica microscope

Posted by Steffen Dietzel on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Zeiss-oil-on-Leica-microscope-tp7591810p7591818.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Petro,

if you check the Material safety data sheet of the Leica oil, you will
find that the producer is Zeiss. We are refilling the nice 10 ml Leica
bottles with plastic rods from a 500 ml Zeiss 518F for years now, no
problem at all.

I do not think that this is the same for other oils though. We had some
substantial chromatic aberrations in the deep red with Cargile HF. So
for a Leica Microscope I would buy either form Leica or from Zeiss. Then
you get what the objectives were calculated for.

Steffen

Am 10.02.2021 um 10:31 schrieb Avi Jacob:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> My completely unsubstantiated guess is, that all the oil (Zeiss, Leica,
> Olympus) is made by the same manufacturer. They all seem to have the same
> details on them.
> Avi
> --
> Avi Jacob, Ph.D.
> Head of The Kanbar Light Microscopy Unit
> The Goodman Faculty of Life Sciences
> Bar-Ilan University, Ramat-Gan 5290002, Israel
>
>
>
> On Wed, Feb 10, 2021 at 11:20 AM Khoroshyy Petro <[hidden email]>
> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Thanks,
>> A nice joke played with me :). I will pay it back once he comes again :)
>>
>>
>>
>> On Wed, Feb 10, 2021 at 10:09 AM Eva Wegel (JIC) <[hidden email]>
>> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your
>> posting.
>>> *****
>>>
>>> Hi Petro
>>>
>>> We have been doing this for more than six years without any problems on
>>> confocal and widefield systems.
>>>
>>> Best regards
>>> Eva
>>>
>>> Dr. Eva Wegel
>>> JIC Bioimaging
>>> John Innes Centre
>>> Norwich Research Park
>>> Colney Lane
>>> Norwich NR4 7UH
>>>
>>>
>>>
>>>
>>> -----Original Message-----
>>> From: Confocal Microscopy List <[hidden email]> On
>>> Behalf Of Khoroshyy Petro
>>> Sent: 10 February 2021 08:21
>>> To: [hidden email]
>>> Subject: Zeiss oil on Leica microscope
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your
>> posting.
>>> *****
>>>
>>> Hi, list.
>>> I was told recently by a Leica engineer that he encountered many Leica
>>> objectives destroyed by Zeiss oil. I was not able to figure out from him
>> if
>>> he was pooling my leg (which I think is the case) or was talking
>> seriously.
>>> Thanks.
>>> Petro.
>>>
>>
>> --
>> ______________________________
>> Petro Khoroshyy
>>
--
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Biomedical Center (BMC)
Head of the Core Facility Bioimaging

Großhaderner Straße 9
D-82152 Planegg-Martinsried
Germany

http://www.bioimaging.bmc.med.uni-muenchen.de