> We use APC Back UPS Pro sinewave units for a lot of sensitive equipment
> including CCD and CMOS cameras, test equipment, and other science-related
> instruments.
> They have an excellent track record and reliability.
> Eg BR1500VA and other models.
>
https://www.apc.com> Good luck.
> Colin
>
> -----Original Message-----
> From: Confocal Microscopy List <
[hidden email]> On
> Behalf Of Mario Methot
> Sent: February 22, 2021 8:50 AM
> To:
[hidden email]
> Subject: Anyone uses UPS units to protect your confocal/2-P systems?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>