Re: Anyone uses UPS units to protect your confocal/2-P systems?

Posted by Jacqueline Ross on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Anyone-uses-UPS-units-to-protect-your-confocal-2-P-systems-tp7591892p7591901.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mario,

One more thing I forgot to say is that our 20kVA system has an additional use as an alert system. If the power goes off, it is set up to send a signal to a phone relay system which calls our contractors and alerts them to the power outage. We have 4 phone numbers so it calls people in order and if the first 3 don’t answer, I get the call and can alert the necessary staff.

Kind regards,

Jacqui

From: Confocal Microscopy List <[hidden email]> On Behalf Of Jacqueline Ross
Sent: Tuesday, February 23, 2021 12:09 PM
To: [hidden email]
Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?

Caution - Forged Internal Domain!
This e-mail cannot be validated and may not have been sent by the sender shown in the 'From' field.
If you were not expecting to receive this e-mail we recommend you contact the sender to confirm that they sent it.
If you believe this email was legitimately sent, we suggest the sender notify the Staff Service Centre that it has been received as a forged (fake) e-mail.
Please contact the Staff Service Centre on extension 86000 if you require further assistance.

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
Post images on http://www.imgur.com<http://www.imgur.com> and include the link in your posting.
*****

Hi Mario,

We have a UPS for every system in our facility. They are primarily 3kVA systems, with a couple higher spec, e.g. the one for 3 phase (20kVA – over spec but was the only one available at the time).. They are mainly Socomec but we also have MGE. They are all online UPS so smooth as well as protect from spikes/outages.

Every microscope user is trained to listen for the beeping and shut the system down if this is ongoing. Also, we have them all plugged into essential power sockets. These sockets are provided with power by generators in the event of a complete campus power failure but don’t come on if it’s just local, e.g. one building/floor.

We started adding these to every new system after having had a very expensive repair bill ($10k) after our Leica confocal went off and on again. They are included in our business cases as a necessary expense.

The UPS are cheaper than the necessary repairs. We also have a contract with a local supplier to check them and replace batteries regularly,. Otherwise, they wouldn’t be much use.

They are also really good for the time-lapse systems because a short outage or spike would otherwise terminate the experiment immediately.

Kind regards,

Jacqui

From: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> On Behalf Of Mario Methot
Sent: Tuesday, February 23, 2021 2:50 AM
To: [hidden email]<mailto:[hidden email]>
Subject: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy><http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>
Post images on http://www.imgur.com<http://www.imgur.com><http://www.imgur.com<http://www.imgur.com>> and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario