Hi everyone

Nice discussion. So if I summarize what is needed:
- good UPS with batteries regularly checked and replaced when needed
- back up power from a diesel generator that takes a short while to take over (basically it is not enough to have the back up power without UPS)
- someone on call who can come before the generator has had enough

This is exactly the conclusion we came to when designing our new building.
But all this is necessary only if there are power cuts several times per year.
If the power cuts come seldom and there is a good surge protection system, there is no need for these expenses. This is of course to be balanced with the cost (and number) of the wasted experiments.

Since we have less than one power cut per year, we have skipped UPS, backup power and on call person and focussed on the surge system.

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Blickagången 16,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI website
Follow our microscopy blog!

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Zdenek Svindrych
Sent: 23 February 2021 03:52
To: [hidden email]
Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=MG41NQPGWIwA4pSzr4INZy8ITCv34%2FuP2S0VebKkSBQ%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=z33rkZt1nl8CkIqzUpW42Ko3GBKPYxNau%2BcU5IuxxTM%3D&amp;reserved=0 and include the link in your posting.
*****

Hi Mario,
I prefer "online" (or double conversion) style UPS.
One 2P system I used to work with was backed up by a massive 6 kW GE Digital Energy UPS (maybe an overkill, but there was an Ar-ion laser as part of the system). Some smaller GE Digital Energy units (1 - 2 kW) were also sold as Franek. Very robust, but rather old.
We also use GE VH-series units on some smaller equipment, like plate readers and liquid chromatography systems. I now only have one microscope backed up, it's a spinning disk confocal (no Ar-ion laser), and the UPS is APC SURTA2200 (it's the "RT" series, that means "online"). I got it for
$200 and I'm happy with it.
Of note is that the microscope is connected to emergency power, and there are more outages on emergency power than on regular outlets (once or twice a year to test the generator and switches, cca 15 second outage to crank up the diesel engine). So for any computerized equipment connected to the emergency power a UPS is a must!
Best, zdenek

On Mon, Feb 22, 2021 at 9:00 AM Mario Methot <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists
> .umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cs
> ylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1
> cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFp
> bGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn
> 0%3D%7C1000&amp;sdata=MG41NQPGWIwA4pSzr4INZy8ITCv34%2FuP2S0VebKkSBQ%3D
> &amp;reserved=0 Post images on
> https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=z33rkZt1nl8CkIqzUpW42Ko3GBKPYxNau%2BcU5IuxxTM%3D&amp;reserved=0 and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building)
> and our systems (and particularly lasers) don't like it too much... I
> was wondering if some of you are using UPS units to protect your
> confocal/2-P systems? If so what type of UPS are you using? All the
> info you could provide would be most welcome, thanks in advance.
>