> Hi Michal,
> you may want to use much longer exposure times (30 seconds for CMOS
> cameras, hours for CCDs) depending how weak your luminescence is. With
> EMCCDs, since the readout noise is negligible, you should get the same
> results by summing shorter exposure images, like you suggested. You may
> also bin the pixels for more SNR.
> For optimization and troubleshooting you can measure the luminescence from
> your cells with a suitable plate reader, they are much more sensitive.
> I haven't tried this with the Zeiss Axio stand, but one issue I found with
> the Nikon Ti-E and TE2000 is the stray light coming from internal
> components (probably the encoders). I had to power off the microscope body
> completely to get low background (I think Ti2-E can power off the encoders
> only, if needed).
> Also, dedicated systems may have much better light throughput, especially
> when you trade off resolution for collection efficiency (just compare the
> back focal plane aperture size of your microscope objective lens with let's
> say Nikkor 50 mm F/# 1.2.
> Good luck!
> Best, zdenek
>
> On Mon, Mar 8, 2021 at 8:50 AM Michał Majkowski <
> [hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Dear List Members,
> > I am just thinking about the use of our Elyra 7 system to visualise
> > bioluminescence.
> > Have any of you ever tried something similar? In systems dedicated for
> > bioluminescence signal is detected with the same or similar camera (Andor
> > iXon Ultra 897) though light path is less complex.
> > My idea is to take time series with longest possible exposure time (1
> sec),
> > convert T->Z and then sum up all slices. Does it have any sense?
> > Thank you in advance