Re: water condensation on camera sensor

Posted by VerĂ³nica Labrador on
URL: http://confocal-microscopy-list.275.s1.nabble.com/water-condensation-on-camera-sensor-tp7591995p7591998.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

We have the same problem. Our Zyla 5.5 camera is coupled to a Zeiss 200M microscope. Acquistion is run by MicroManager 1.4.22 ImageJ. We solve this condensation problem deactivating the Camera Sensor Cooling option from Tools>Device Property Browser... everytime we use the system.

Hope this helps if you are in the same situation.

Kind regards,

VerĂ³nica