Re: water condensation on camera sensor

Posted by Cammer, Michael-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/water-condensation-on-camera-sensor-tp7591995p7591999.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Sorry to hear you have this problem. This exact question has reoccurred on this list over the years.  This happened to our Zyla 5.5 and other people have reported similar.  If the seal breaks, water condenses on the chip or inside front plate when the cooling is on.  Andor can fix this, but they may charge you.

Regards-

Michael Cammer, Sr Research Scientist, DART Microscopy Laboratory
NYU Langone Health, 540 First Avenue, SK2 Microscopy Suite, New York, NY  10016
Office: 646-501-0567 Cell: 914-309-3270  [hidden email]<mailto:[hidden email]>
http://nyulmc.org/micros  http://microscopynotes.com/
Acknowledgement in your publications and presentations of work performed in the Microscopy Core plays a vital role in securing support and the funding necessary to maintain and operate this valuable research resource.   For publications that were made possible by work performed in the core, please use the acknowledgement statement "We thank the NYU Langone Microscopy Core for experimental and technical support" and include required grant numbers as listed here http://microscopynotes.com/ilabnyu/acknowledgements2017.pdf
Please also consider staff for co-authorship if they played a key role in the study.



________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Edite Figueiras <[hidden email]>
Sent: Wednesday, March 17, 2021 7:50:46 AM
To: [hidden email]
Subject: water condensation on camera sensor

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear All,

I wonder if anybody in the list had the same/similar issue we have in the
lab.

We have water condensation on the camera sensor (Andor Zyla camera) when we
turn on. It disappears when we turn off the camera.
The camera is attached to a lens as shown here:
https://drive.google.com/file/d/1aZLiSj4wI7xEzDaSsrVewVajyYMqtUcu/view?usp=sharing

This is how it looks in the image:
https://drive.google.com/file/d/1F6NUd4X6VmbCy77GhrppxMGXZlKFDjqA/view?usp=sharing

Do you have suggestions on how to fix this?

Thank you for your help!
Best wishes,
Edite

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================