Are lower magnification objectives brighter?
Posted by
0000001ed7f52e4a-dmarc-request on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Are-lower-magnification-objectives-brighter-tp7592013.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Dear all,
Are lower magnification objectives brighter than higher magnification ones when they have the same NA, e.g. a 40x NA 1.4 objective compared to 63x NA 1.4? I mean for confocal microscopy.
Confocal.nl stated this is a recent webinar and on their website:
“A lower magnification allows for a larger field of view and brighter images, since light intensity is inversely proportional to the magnification squared”
https://www.confocal.nl/#rcm2I would think that this is caused by less light going through the smaller back focal aperture when the illumination is held constant? Most of the light is clipped as explained in fig 1 of
https://www.nature.com/articles/s41596-020-0313-9So, the microscope manufacturer could adjust the illumination beam path and laser powers to best suit the objective?Or are lower magnification objectives really brighter?
The field of view will obviously be larger for the 40x objective, but I am more interested to understand the claimed benefit in brightness.
best wishes
Andreas