Hi Andreas,
If you divide the same amount of light across a more magnified PSF, then
the PSF covers more pixels and so each pixel gets fewer photons. However,
in this case you would also be more densely sampled, and you could
digitally downsample the image, which would have the effect of putting the
same number photons into fewer pixels. If dark and read noise are low,
this would effectively give you the same image as you would have gotten
using a lower magnification to begin with.
Mike